Sam's Educational Michelson Homodyne Interferometer Project Manual V2.1

Assembly and Experiments

(Now Also includes specific information for V1.75 setup)

Version 2.23 (10-Jun-2025)

Copyright © 1994-2025
Sam Goldwasser
--- All Rights Reserved ---

For contact info, please see the Sci.Electronics.Repair FAQ Email Links Page.

Reproduction of this document in whole or in part is permitted if both of the following conditions are satisfied:
  1. This notice is included in its entirety at the beginning.
  2. There is no charge except to cover the costs of copying.


Table of Contents


Preface

Author and Copyright

Author: Samuel M. Goldwasser

For contact info, please see the Sci.Electronics.Repair FAQ Email Links Page.

Copyright © 1994-2025
All Rights Reserved

Reproduction of this document in whole or in part is permitted if both of the following conditions are satisfied:

1. This notice is included in its entirety at the beginning.
2. There is no charge except to cover the costs of copying.

DISCLAIMER

The information in this document is intended for use in hobbyist, experimental, research, and other applications where a bug in the hardware, firmware, or software, will not have a significant impact on the future of the Universe or anything else. We will not be responsible for any consequences of such bugs including but not limited to damage to the $100,000,000 wafer FAB that was purchased on eBay for $1.98 + shipping, financial loss from the waste of 28 spools of ABS due to the office 3-D printer fabricating a part with random dimensions due to loss of lock, or bruising to your pet's ego from any number of causes directly or indirectly related to the implementation and use of this system. ;-)

ACKNOWLEDGMENT

The Michelson Interferometer experimental setups V1.0 and V1.5 were originally developed for Engineering student projects at Swarthmore College, Pennsylvania.

SAFETY

The only safety issues with the experiments to be performed using this kit are with respect to the low power Helium-Neon (HeNe) laser. Sure, you could drop the breadboard on your foot, but that's outside our control. :( :-)

Even though this laser is not likely to cause any harm, one should always take laser safety seriously. Someday you may be working with one that is truly dangerous.



  • Back to Sam's Educational Michelson Homodyne Interferometer Project Manual Table of Contents.

    Abstract

    The Michelson interferometer is one of the earliest and simplest to be developed, but also likely the most widely used configuration in a variety of applications including metrology (precision measurement). The type found in most introductory textbooks use single frequency or unstabilized lasers in what are known as "homodyne" interferometers and those are the subject of this manual. Another version uses a special two frequency laser to implement the "heterodyne" interferometer, which are more likely to be found in high performance commercial applications like semiconductor wafer foundaries and are the subject of the companion document: Sam's Educational Michelson Heterodyne Interferometer Project Manual V2.1.

    An experimental setup is presented which allows for several types of interferometers to be easily implemented without requiring any special tools or test equipment. The behavior of various interferometer configurations will be explored as well as how the characteristics of the laser impact performance. Various enhancements are also described for both the laser and detector, as well as extensions to actual measurements like displacement (change in position) down to nm precision. The set of parts may be easily duplicated and/or modified for specific interests.



  • Back to Sam's Educational Michelson Homodyne Interferometer Project Manual Table of Contents.

    Introduction

    IMPORTANT: This manual applies to version 2.1 of Sam's Educational Michelson Homodyne Interferometer Kit and also includes support for V1.75, which differs primarily in the heights of the components and having fewer options. Here are links to all of them:

    For the combined kit (heterodyne + homodyne), both of the relevant manuals will need to be referred to, though there is a lot of overlap.

    Detailed information and instructions on using and constructing most of the sub components of these kits like the various custom PCBs may be found at Sam's Electronics and Laser Kit Information and Manuals. There will also be links to them throughout this manual.

    Homodyne interferometers employ a laser which nominally produces a single optical frequency ("laser line") while heterodyne interferometers employ a laser that generates two closely spaced optical frequencies. They each have their advantages and drawbacks. Much more on this below.

    The photo below shows the Basic version configured for the LI. There are minor variations in the actual kits depending on version and whether they are the "Basic", "Deluxe", or "Deluxe+" versions, though this layout will work for all of them. Differences will be noted below. Where the total path length and PLD is small, no beam expander is needed for the laser as in the Basic version. This may result in a larger signal since the entire beam hits the detector. But the smaller beam makes it more sensitive to alignment.

    Typical Linear Interferometer Setup with Quadrature Decoder and µMD0

    Interferometers are the key technology is numerous applications in manufacturing and testing where the very minute wavelength of light is the "yardstick" providing non-contact measurements down to nanometer precision. In short, a light source is split into two parts that may travel different paths and then recombined at some type of detector. Where the path lengths differ by an integer number of wavelengths, the result will be constructive interference and the output of the detector will be high; where it differs by an integer number of wavelengths plus one half wavelength, the result will be destructive interference and the output of the detector will be low. In between, the output will vary sinusoidally. With suitable detectors and electronics, remarkably precise measurements can be performed. For example, nearly every microchip manufactured in the explored universe has been done with wafer steppers whose stages were positioned using interferometry based on HeNe lasers.

    While interferometers are employed in a wide array of applications, the general emphasis for these experiments relate to the use of interferometery in metrology - precision measuremens of physical characteristics like displacement, velocity, angle, straightness, and more. Therefore unlike numerous interferometer experiments that may be found via a Web search, the emphasis here is on the signals that the setup provides, not so much on the interference patterns. (Though nothing precludes the observation of these.

    The experimental setups will enable various interferometer configuration to be easily implemented and then tested with one arm being on a micrometer linear stage and/or with some other device or material that can vary the path length precisely.

    The light source is a Class II 633 nm Helium-Neon laser (HeNe for short) with an output power of between 0.4 and just over 1 mW. The basic detector is a biased photodiode connected to a dual channel digital oscilloscope. Variations and enhancements to these will be offered as options.

    Among the areas that can be explored with the Basic setups are:

    There is no need to construct all of the interferometer configurations described below. Doing the Linear Interferometer (LI) first makes sense since there are detailed instructions on its construction, alignment, testing. Building the High Stability Plane Mirror Interferometer (HSPMI) would be the logical next step moving from cube corners to plane mirrors. It also permits the loudspeaker and/or PZT actuators to be added. Then after that one of the others. Perhaps coordinate with the other project students using this same kit so that each of you do different ones.

    This minimal set of experiments can all be done using parts in the Basic Kit:

    The following additional projects can be done using parts in the Deluxe kit:

    The following are more advanced projects, but they may require additional parts and/or different parts including the laser that are not included in the either kit:

    There is some information on these in this manual and links will be provided to learn more.

    As of Winter 2022, there are 3 versions of the setups. Around 5 each of V1.0 and V1.5 (which differ in minor details) have been built and are being used for in-person and remote project labs at a local college; V2.1 is the one for sale going forward and comes in several flavors. The detailed asseembly instructions in this manual are for V2.1.

    Various configurations of these kits have been available on eBay under my user ID: siliconsam. But they have been discontinued due to infrequent sales resulting in each being a one-off and more effort than was worthwhile. However, if you are really interested - and motivated - it may be possible to work out some cooperative arrangement directly. There is a complete parts list at the end of this manual. Most opto-mechanical parts are readily available from Thorlabs and the hardware from McMaster-Carr. I can provide the laser at a relatively affordable price. The custom parts can be relatively easily fabricated, or I may be able to provide them at modest cost. Contact me via the Sci.Electronics.Repair FAQ Email Links Page if interested in pursuing this.

    Some of the photos here are of the original prototype on a custom aluminum optical breadboard. They are only for reference (and because I'm too lazy to reshoot them!).

    Note: Off-page links (including any clickable graphics) open in a single new window or tab depending on your browser's settings. A suitable fixed width or monospace font like "Courier New" must be specified in your browser to make sense of the simple ASCII diagrams. For Firefox, go to: "Settings", "General", "Fonts", "Advanced", "Monospace", and confirm that it is "Default (Courier New)".



  • Back to Sam's Educational Michelson Homodyne Interferometer Project Manual Table of Contents.

    Interferometers for Metrology Applications

    Classic Michelson Interferometer

    All of the measurements performed by these systems are based on variations of the Michelson interferometer, invented over 100 years ago by Albert Michelson. This is one of the simplest interferometer configurations but also one of the most widely used. Even though a two frequency laser is used here instead of the linearly polarized laser in the other kits, most of the basic priciples are the same. The specific differences will be noted.

    The textbook version is shown below:

    In short, a light beam is split into two parts which are bounced off of a pair of reflectors and recombined at a detector. Any change in the relative path lengths of the two "arms" formed by the reflectors results in a phase shift between the waves in the two beams resulting in constructive or destructive interference, which can be measured and converted to displacement (change in position) down to nanometer precision. All other types of measurements made by these systems are based on opto-mechanical configurations designed such that changes in the measured variable are detectable by what is in essense a Michelson interferometer.

    Where the Path Length Difference (PLD) between the two arms is small, the requirements for the laser are not very stringent. In fact, for very small PLDs, an LED or even a totally incoherent source like an incandescent lamp may be substituted for the laser. However, to be useful for the PLDs necessary for most applications (millimeters to 10s of meters or more), the light source must be a laser. And not just any laser, but one that has a narrow "linewidth". While the popular concept of a laser is of a light source that is monochromatic (single color or wavelength), in reality most lasers do not even come close. It takes careful design and implementation to achieve that. For these metrology applications using the so-called "homodyne" interferometer, the laser should ideally produce an output that is a single optical frequency with a linewidth approaching zero. In practice, it isn't that narrow but can still result in a linewidth of much less than 1 MHz, resulting in a usable PLD of 100s of meters. This is usually a low power stabilized HeNe laser. However, for PLDs of a few inches, a common (much less expensive) unstabilized HeNe laser will be adequate. Within this range, all the modes will be sufficiently correlated and the fringe contrast/signal amplitude will be acceptable.

    For the heterodyne interferometer, instead of a single optical frequency, there are two spaced by up to 20 MHz. The usable PLD is determined by this "split" frequency and is thus still very large. The split frequency only limits the maximum slew rate of the remote reflector. More on this later.

    Engrave PLD on your brain. It will be used throughout this manual. :)

    Cube Corner Retroreflectors

    The simple Michelson interferometer setup can be used in a metrology system, but it has severe limitations which make it impractical for most applications. Alignment is extremely critical. Even the slightest deviation from perfect alignment will result in a reduction or loss of signal. Yet when perfectly aligned, one half of the optical power from the laser reflects directly back into the laser - which may destabilize it resulting in erratic fluctuations of its output in amplitude, optical phase, optical frequency, and polarization. HeNe lasers will not be damaged, but this is not true of some other types of lasers.)

    Near the end of this manual is an experimental setup similar to the original Michelson Interferometer configuration adapted for the Polarizing Beam-Splitter. So you can experience these issues for yourself.

    The diagrams below show some variations on the angles of the Beam-Splitter (BS) and Mirrors (M1, M2).

    The first one is essentially the same as the diagram, above. The "Single Angled" one does nothing to prevent back-reflections. And while the "Double Angled" version does eliminate back-reflections, the two beams are angled at the detector which means their relative phase will vary across the detector. That is undesirable for our experiments in at least two ways: First, it will mean that the detector won't see a clean fringe signal. In fact there may be little or no signal depending on how the fringes average across its face. Second, alignment will change as either mirror is moved making testing with respect to PLD very tricky. But can you suggest an application where it may be useful?

    The first enhancement of the Michelson interferometer is to add a means of separating the outgoing and return beams so that there is vitrually no optical power returned to the laser. The simplest way to do this is to replace the mirrors with Retro-Reflectors (RRs), typically cube-corner (trihedral) prisms, which have the property of returning the beam directly back parallel with the outgoing beam, but which may have an offset. In this way, virtually none of the reflected light ends up back at the laser. The use of the RRs also greatly reduces the sensitivity to alignment as any change in their angle is converted to a small change in the distance between the outgoing and return beams, but they remain parallel.

    All of the practical interferometer configurations include at least one cube corner retro-reflector.

    Polarizing Beam-Splitter

    The second enhancement is to use a Polarizing Beam-Splitter (PBS) in place of the 45 degree partially reflecting mirror. While a plate beam-splitter could be used, the PBS cube (or simply PBSC) is much more common. (The acronyms PBS and PBSC will be used interchangeably throughout this document.)

    The beams reflected to the two arms of the interferometer then have orthogonal polarization which effectively makes them independent until they are combined at the detectors.

    The result is then one of the most widely used configurations - the Linear Interferometer (LI), which was the first one used by HP in their original 5500A laser interferometer displacement measuring system. (As an aside, I do NOT know where that name "Linear Interferometer" comes from except perhaps that the inital configuration was in-line with the laser and thus "linear".) In practice, Arm 1 is used as the reference and is made as short as possible with the Cube Corner (CC) attached directly to the PBS cube. Both arms can move where differential measurements are required.

    With the use of the PBS, the maximum amount of the laser optical power is available - virtually nothing exits out the unused side of the beam-splitter as it would in the original Michelson setup, above. However, at least half the power is lost in the detection scheme that is typically used so it would end up being similar. In principle, this wasted power can be diverted to a second detector. Their difference will then have twice the amplitude and the signal-to-noise ratio will nearly double (but with opposite phase). This is rarely, if ever, done though. But for the more complex multi-pass interferometer configurations described later as well as for use with two frequency lasers, the use of the PBS is essential to avoid incurring very large losses, or for the schemes to work at all.

    As noted for the heterodyne interferometer - probably the most popular type used for high performance metrology applications and the topic of this kit - the laser actually produces a pair of optical frequencies that are othogonally polarized. They are typically separated by 100s of kHz to 10s of MHz called the "split" or "REF" or reference frequency. Most commonly the split frequency is in the small number of MHz range. For the HP/Agilent 5517 laser in this kit, it has been adjusted to be between 1.5 and 2.0 MHz. F1 is the lower optical frequency and is the horizontally polarized component; F2 is the higher optical frequency and is the vertically polarized component. As long as the difference frequency between F1 and F2 is reasonably stable, the specific value only impacts the maximum slew rate possible for the moving mirror or cube corner. It has no effect on the resolution. Think of the split frequency as a carrier that can be conveniently manipulated by AC electronic circuits. So just as with radio, the carrier does not carry information but makes it convenient to process that information. There are also several other advantages including reduced affects of signal amplitude and less sensitivity to interferometer alignment. Everything with respect to the interferometer optics and configurations is similar except that instead of the phase change between the two (essentially base-band) polarized components derived from the laser, it is the phase change between the REF signal (F2-F1) and a "MEAS" signal (F2-F1+ΔF1) with ΔF1 being a result of the Doppler shift from the moving reflector.

    The maximum PLD is virtually unlimited for the two frequency laser with usable PLDs of 10s or even 100s of meters or more - impacted more by issues of beam divergence and collimation than the stability of the laser itself. However, since it is the PLD that is being measured, you can't purge it from your brain, sorry. ;-)

    Common Interferometer Configurations for Displacement Measurements

    Here are several interferometers that may be used for measuring displacement (change in position). The first 4 are the most common.

    When used in a displacement measuring system with a 633 nm HeNe laser, 1X represents a full cycle resolution of 1/2 wavelength or ~316 nm; 2X of ~158 nm, and 4X of ~79 nm. For a homodyne system with a Quad-A-B deterctor, there are 4 counts per cycle so that gets multiplied by up to 4. For a heterodyne system using µMD2, there are 2 counts per cycle so that gets multiplied by up to 2. The very commonly used PMI or HSPMI will then have a heterodyne resolution of ~79 nm and interpolation techniques can extend it down to under 1 nm. However, if the signals are noisy or asymmetric, the benefits may not be achievable.



  • Back to Sam's Educational Michelson Homodyne Interferometer Project Manual Table of Contents.

    Detectors - Optical to Electrical Conversion

    With the emphasis on the signal behavior of interferometers, the conversion from light to electrical is critical - but very simple to implement. To be able to view or do anything with the output of the interferometer requires some means of converting light to an electrical signal. The type of detector use here is called a "biased photodiode" and is essentially a current source with an output proportional to incident optical power.

    Biased Photodiode (BPD1)

    A silicon photodiode (PD) when reverse biased by a positive DC voltage (battery or power supply) allows a current (designated Ipd) to flow with a sensitivity measured in amps / watt (A/W) or for our purposes, mA/mW of incident laser power. The sensitivity for silicon at 633 nm is typically between 0.3 and 0.4 mA/mW and linear up to several mW for the types of PDs used here. This relationship holds even when a load resistor R-Load is installed between the PD and circuit common (negative of the bias supply), resulting in an output voltage that is linear with respect to incident laser power based on Ipd * R-Load. For example, with a laser power of 0.5 mW, a PD sensitivity of 0.35 mA/mW, and R-Load of 10K ohms, Vo would be 1.75 V.

    The most basic circuit is shown below:

               Silicon 
              Photodiode
       +---------|<|-------+-------o Output to scope or DMM
       |  Cathode   Anode  |
       |                   /
       |                   \ R-Load
       |                   /
       |     Bias Supply   \
       |       +| | -      |
       +--------||||-------+-------o GND / Common / Return
                | |
    

    Note the polarity of the PD with its cathode connected to the positive of the power supply and thus reverse biased. With no light incident on the PD, only the so-called "dark current" will flow, which is generally small enough to be ignored (nanoamps or less). Actual usable detectors are almost identical to this simple circuit.

    Single Channel Detectors

    Here is a widely used commercial biased photodiode detector and a version used in the kits:

    Testing of either detector can be done using the laser or even a flashlight to confirm sensitivity to light. However, even a super-bright flashlight will likely result in only a small signal compared to the laser.

    Dual Channel Detectors

    These only use the custom detector. It would be silly, space consuming - and expensive - to require two DET110s! So, for two channels, all parts except for the power source and protection resistor are replicated:

           R-Protect    PD1      Yellow
       +-----/\/\----+--|<|---<<-----------------------+----------o Scope Channel 1
       |             |                                 |
       |             |  PD2      Blue                  |
       |             +--|<|---<<---------------------------+------o Scope Channel 2
       |                                               |   |
       |                                               /   /
       |                                       R-Load1 \   \ R-Load2
       |                                               /   /
       |                                DC Power       \   \
       |                          Red    +| |-   Black |   |
       +-----------------------<<---------||||---------+---+------o Scope Ground
                                          | |
    
       |<--- SBB or QDx PCB --->|<----- Scope / Power Wiring ------>
      

    PD Pins: Facing Front of PD with Legs Down: Anode on left, Cathode on right.

    R-Protect: 250-1K ohms typical. R-Load: 10K-1M ohms typical.

    The same PCBs may be used for this circuit.

    What is done with the light before it hits the photodiodes is the interesting part, which will be described where relevant. ;-)

    Dual Channel Quad A B Preamp (AB2)

    The problem with the simple circuits above is that there is no amplification. So to achieve both a usable signal level and reasonable bandwidth is not possible. By using a load resistor with a high enough value, it the signal level will be acceptable but then the badnwidth may as low as a few kHz. While this is satisfactory for initial testing and many of the experiments, greater bandwidth will eventually be desired for the displacement readout.

    The purpose of the Quad-A-B Preamp (henceforth referred to as QAB2 or simply AB2) is to provide a simple solution that accepts photodiode inputs and generates differential RS422 A and B signals that can be input to µMD0, µMD1, µMD2, or another compatible displacement measuring system.

    QAB2 is on a 1.6 inch by 2.25 inch PCB and runs on 12 to 15 VDC. (The PCB itself is called SG-AB2.) The optical input is a beam up to ~3 mm in diameter (using the default photodiode) with an optical power from <25 µW to >1 mW. QAB2 has >3 MHz bandwidth (full cycle) which is more than adequate for systems using the kit lasers as well as for many real applications. With a Linear Interferometer which has a full cycle of ~316 nm, the slew rate can be greater than 1 meter per second, which is a fairly nutty velocity. ;-). And it is expected that the bandwidth limit can be extended with trivial changes to only a few part values. This is left as an exercise for the student. ;-)

    Populated AB2 V1.00 PCB

    Although shown with photodiodes plugged into the PCB, in actual use, they would probably be on the detector PCB with short twisted wire cables to connect them to the SG-AB2 PCB.

    Parts to construct AB2 will eventually be included as part of the homodyne setups. More information on AB2 including complete "Heathkit™-style" assembly instructions may be found at Quad-A-B Preamp 2 (QAB2) Assembly and Operation Manual.



  • Back to Sam's Educational Michelson Homodyne Interferometer Project Manual Table of Contents.

    Linear Interferometer

    Although HP called the combination of the PBSC and reference CC attached to it, the "Linear Interferometer", the term "LI" here will refer to the entire setup. And to enable the PLD to be set to zero, the CC normally attached directly to the PBSC will be mounted a few inches away.

    The designations m-n show the paths taken by the Arm 1 and Arm 2 beams where "m" is the Arm and "n" is the sequence number.

    Linear Interferometer Setup

    The diagram below shows the general arrangement of the laser, beam splitter, cube corners, and detector.


    Linear Interferometer Diagram and Suggested Breadboard Layout

    The other configurations will have a few additional or substitute parts and small variations in the horizonatal position of the laser and placement of the detector but are otherwise similar. Therefore the LI setup will be described in more detail.

    A variety of mounting schemes are used:

    Setting the Heights

    The following diagram shows the relationships of the various mounts for setting the heights during assembly. This diagram applies to all V2.1 setups. 4-1/4" places the laser in the center of the PBS vertically:


    Required Optical Components and their Suggested Heights - V2.1

    And this diagram applies to all V1.75 setups. 3-1/4" places the laser in the center of the PBS vertically:


    Required Optical Components and their Suggested Heights - V1.75

    Note that the primary differences between these are in the size of the posts and postholders as V1.75 does not need to accomodate the additional height of the optional motion control components. Similarly, everything in V2.1 can be reduced in height by around 1" if there is no desire to accomodate them. V1.75 also assumes the use of the Parker 3902 stage which has a thickness of 1/2". To use the typical generic stage requires that everything be raised by a bit over 1/4". This isn't rocket science, you'll figure it out. ;-)

    The heights of any Retro-Reflectors (RRs) in the setup will be what most affect the beam height. This is true of the Linear Interferometer (LI). Where there is an RR attached to the PBS cube like the Plane Mirror Interferometer (PMI), the alignment of the laser will need to be used. However, there is a wide tolerance and enough degrees of freedom so in the end, it should not really be much of a problem to set it up.

    Here is an annotated photo of the Deluxe version configured for the HSPMI:


    Deluxe Version Configured for the High Stability Plane Mirror Interferometer with µMD0

    And if you have a sharp eye, you will notice that this is actually not exactly a V2.1 or V1.75 setup but should be close enough for government work. Live with it. ;-)

    The Deluxe+ version would be similar except for the larger breadboard and longer rail; The Basic version does not include the beam expander, mounted CC, QWPs, or planar mirrors.

    Assembly and Alignment of the Linear Interferometer

    The following procedures may be used to install each of the parts and then get to the point of "first signal" using the oscilloscope. The assembly for the other interferometers will be virtually identical in most respects, but may be trickier depending on the type.

    It is assumed that nothing has been mounted, but depending on the previous use, some of these steps have already been completed. Refer to the layout diagrams, above, parts locations that are known to work.

    Parts attached with fasteners should be snug but don't overtighten.

    It is also assumed that the laser is linearly polarized. Slight changes are required if it is random polarized.

    Please refer to the appropriate Heights diagram. Clicking on it will open a new window, or print it out for reference.

    1. Laser mount posts: Attach two posts to the breadboard using 1/4-20 set screws. Tighten securely.

      If the laser head has a beam expander, or it is anticipated that one will be added later, it is recommended that the posts be mounted 1 hole to the left of where the are in the photo.

      Note: To assure that there are ample threads engaged in both parts here and in subsequent steps with a post attached directly to the breadboardd or mounting plate, the set-screw should be installed approximately half-way and then a thin tool or the edge of a piece of stiff cardboard can be used to keep the set-screw from turning as the post or post holder is threaded onto it before tightening.

    2. Laser head: These steps secure and align the laser head cylinder.

      • Install the Nylon thumbscrews in the ring mounts if not already present. If the thumbscrews are much longer than needed, they can be trimmed. However, since not all interferometer configurations have the laser head in the same position, make sure not to cut them too short!

      • Mounting ring assmebly:

        • New style Small Mounting Rings (thumbscrews at 0/90 degrees with Offset Plate):

          • Attach the rings to the Offset Plates with a pair of 8-32 x 1/2" cap-head screws in the counterbored holes (so their heads do not protrude into the laser space). Use a square or something similar to assure that the rings are perpendicular to the plate. It is recommended that they be attached and then mounted opposing (i.e., rings face-to-face) so that any errors in angle will be the same. Tighten securely.

          • Attach the laser mount rings via the Offset Plate to the top of the posts with 8-32 cap-head screws. For the most common configurations, this should use the front-most hole of the Offset Plate so the beam is 1/4" towards the rear of the breadboard.

          • Slip the laser head into the rings. Adjust the thrumbscrews so it is approximately centered horizontally and vertically.

        • Old style Small Mounting Rings (thumbscrews at ±45 degrees):

          • Install the Nylon thumbscrews in the ring mounts if not already present. If there are two different lengths of thumbscrews (e.g., 1 and 1-1/4 inch), the longer ones should be in the front side of the rings. If the thumbscrews are much longer than needed, they can be trimmed. However, since not all interferometer configurations have the laser head in the same position, make sure not to cut them too short!

          • Attach the laser mount rings to the top of the posts including spacers if necessary. Secure them with 8-32 cap-head screws. For the most common configurations.

          • Slip the laser head into the rings. Initially set it to approximately centered horizontally and vertically. Then shift it 1/4 inch toward the back of the baseplate.

      • Plug the big white male "Alden" connector of the laser head into the female Alden connector of the power supply. Make sure it is seated fully, usually against the shoulder with no part of the prongs visible.

        • If the power supply is "lab style" - a chessy plastic box with keylock switch and power indicator - then it plugs into the AC line. An IEC type line cord will be included if in the USA; you will have to provide the approriate like cord if elsewhere. If there is a voltage switch on the back of the supply, confirm that it is set corretly for your location. If set wrong, the power supply may be damaged.

        • If the power supply is a bare "brick", it requires 12, 15, or 24 VDC from a matching wall adapter that should be included. Confirm that the voltage ratings match. There may be more than one wall adapter in the kit that appear similar. Check the labels! Using the wrong one may damage the brick and/or wall adapter. The wiring is probably already done via a barrel to screw terminal adapter, but confirm that the wires are secure and attached to the correct terminals. Positive is red and negative is black. If there is a yellow wire, it should be attached to the black wire and thus negative as well.

      • Power up the laser by turning the keylock switch or plugging in the wall adapter. It may take a few seconds or even up to a minute to light. If it doesn't light in a reaosnable length of time, check the wiring. Shining a flashlight in the front of the laser will sometimes help it along. DO NOT stare into the beam with your remaining good eye. :-)

      • Confirm that the laser is polarized: The model number should be 1107P, 1108P, or 05-LHP-211. There should be an alignment mark near the front of the laser. If not or just to confirm, rotate an LP or LP/CP in front of the laser. If the laser is polarized, there will be orientations where virtually no light gets through at all times. The polarization axis is then at 90 degrees to this. Add an alignment mark if there is none. (If it is random polarized, this orientation will swap by 90 degrees over a few seconds as the laser warms up. More on this below in the section on linear versus randon polarized lasers.) DO NOT assume the yellow arrow of the safety sticker (if present) identifies the polarization orientation (though it probably does if there is no other mark). But check it anyhow!

      • Orient the laser head so the polarization axis is at -45 degrees (counterclockwise) from vertical when viewed from the front. What differences will there be if it is at +45 degrees?

      • Using the thumbscrews, adjust the position so that the beam is at a height of exactly 4-3/4 inches and perfectly horizontal, and 1/4 inch toward the back of the setup. DO NOT overtighten the thumbscrews - just enough so the laser head won't move on its own.

        Doing this accurately is critical to the ease with which the subsequent alignnment can be performed since not all mounts have sufficient degrees of freedom to accomodate an arbitrary beam location and direction. Fabricating an "alignment aid" out of carboard may be useful. This would have a hole at the optimal height offest 1/4" toward one side from a mark at the bottom.

    3. Interferometer assembly: This consists of the Polarizing Beam-splitter cube, mounting plate, 2x posts spacers, 2x 1/4-20 setscrews, 2x spacers, and 2x 8-32 cap-head screws.

      1. Attach a pair of posts to the breadboard with 1/4-20 x 1/2" setscrews in the locations shown above. Thread the setscrews about half-way into the breadboard. Then with something thin pressing on their sides to prevent them from turning, thread the posts onto the setscrews until they contact the breadboard. Use a thin rod (e.g., hex wrench) through the hole in the post to tighten them.

      2. Attach the "PBS Mount Adapter Plate" to the posts with a pair of 8-32 cap-head screws with spacers between the plate and post. The screw on the right should be made fairly tight as it won't be accessible once the PBSC is installed.

      3. Remove any optics that may be attached to the PBSC. They will not be required for the LI. Store them wrapped in soft paper towels, bubble wrap, etc., to protect the optical surfaces.

      4. Place the PBSC on the PBS Mount Adapter Plate so that the diagnonal marking is front-left to back-right with the "In" arrow pointing left-to-right.

      5. Use four 4-40 x 1-3/4" cap-head screws to secure the PBSC to the plate from the top if the adapter plate holes are tapped or four 6-32 x 1/2" cap-head screws from the bottom if they not tapped.

      The laser beam should pass through the PBSC centered vertically and 1/4 inch toward the back. It should be at the same location relative to the breadboard at the far end. If not, fine tune alignment. :) Getting this dialed in precisely will greatly simplifiy alignment later.

    4. Turning mirror: Use tape or a *tiny* bit of adhesive to attach the turning mirror to the right angle "Turning Mirror Bracket" centered vertically with respect to the PBSC and flush with the right edge (which will be closest to the back of the breadboard). Attach the bracket to the PBS Mount Adapter Plate with a 3/8" 4-40 cap-head screw and #4 washer. Center the screw in the elongated hole for now.

    5. Arm 2: These steps assemble the components of Arm 2. Arm 2 is assembled first because the lateral position of the CC cannot be adjusted relative to the laser.

      1. Thorlabs rail, ball bearing rail, or Motion Control Platform (MCP): This assumes an initial installation, not swap.

        • Thorlabs rail (Homodyne only):

          1. Attach the Thorlabs RLA0600 6 inch rail (Basic and Deluxe) or Thorlabs RLA1200 12 inch rail (Deluxe+ version) to the breadboard. The recommended location is shown in the diagram above.

          2. Add 8-32 screws and #10 washers as high as possible to the ends of the rail to act as stops.

          3. Attach the linear stage to the RC1 carrier using a 4-40 cap-head screw and washer.

          4. Install the stage/carrier assembly onto the RLA rail.

        • Ball bearing rail (Heterodyne or combined):

          1. Attach the MGN15-150mm rail with MGN15C carriage to the breadboard. Try to avoid removing the carriage from the rail as it is then easy for some ball bearings to attempt to escape. :( ;-) Since the holes in the MGN15-150mm rail do not line up with breadboard holes, it is attached to the breadboard with either a pair of 6-32 cap-head screws installed in special tapped holes, or one of these along with a 6-32 screw with nut and lockwasher through a 1/4-20 hole. The recommended location is shown in the diagram above.

          2. Attach the linear stage adapter plate to the MGN15C carrier using four M3-6mm flat-head screws. Tighten these securely as they won't be accessible after the other parts are installed.

          3. Attach the linear stage to the MGN15 adapter plate using four 4-40 1/4 inch cap-head screws at the corners.

          4. Add stops at both ends of the rail using the rubber bumpers (150 mm rail) or screws, spacers, and nuts (300 mm rail) provided. ;-)

        • Motion Control Platforms: These instructions will differ slightly depending on whether the MCP is Type 1 or TYpe 2, and on the length.

          1. If purchased at the same time as the interferometer, holes will already have been drilled in the breadboard for the stepper motor rig, with the motor on the right. These locations will enable the maximum usable travel. Else holes wlil need to be added.

            Secure the platform on the breadboard using four 4-40 1/2" cap-head screws if the holes are tapped, or 4-40 3/4" screws and nuts if not. Take care to assure that it is aligned.

          2. Attach the linear stage adapter plate to the platform using four M3-6 flat-head screws. To maximize travel, the plate should extend to the left (toward the PBS).

          3. Attach the linear stage to the linear stage adapter plate using four 4-40 1/4 inch cap-head screws at the corners.

          For more on setting up and using the MCP, see the section: Motion Control Platform

      2. Refer to the "Suggested Heights" diagram and attach a 2" post holder (rail or slide) or 1" post holder (MCPs) to the linear stage with an M3-12mm cap-head screws and #4 washer through the hole in the bottom of the post holder. Orient it so the knob on the post holder is easily accessible. Center the screw as best as possible and tighten it securely. Note: The hole may be tapped 4-40, in which case the screw should be 4-40 1/2".

        CAUTION: Make sure the tip of the screw does NOT contact the fixed part of the linear stage. If it does, an addition washer may need to be added. Or the screw can be shortened slightly with a metal file or grinding wheel.

      3. Adjust the micrometer so the stage is approximately in the center of its travel range.

      4. KM100 or similar mirror mount: Secure to a 1" post (rail or slide or MCP Type 1) or 3/4" post (MCP Type 2) with an 8-32 3/8" cap-head screw. Slip the post into the Arm 2 post holder and hand tighten its thumbscrew. Adjust the two alignment knobs so that the mounting plate is parallel to the base in both directions. If the adjustments are too tight, check that any locking screws are not tightened. Else, total removal, cleaning, and lubricating with a tiny amount of light grease or machine oil will be required.

      5. Cube Corner (CC) trihedral prism: Install the CC in the KM100 with its apex facing out and oriented to that a flat is at the top or bottom. (This reduces the chance of the beams hitting an edge of the prism.) It should be secured with either a soft-tipped set-screw or Nylon wide-head screw. DO NOT overtighten - it should be snug enough not to fall out (these are fragile!) but not so tight as to smash the CC! Note that the CCs mount backwards from what might be expected so that their edge can be secured properly. It's flat surface is facing through the mount.

      6. Set the knobs of the adjustable mount so its fixed and moving plates are approximately parallel and orthogonal to the beam axis as in the diagrams.

      7. The horizontal and vertical position (not orientation) of the CC relative to the incident beam adjusts the corresponding coarse position of the return beam.

        • Since the adjustable mount on Arm 2 and thus its CC cannot be moved laterally relative to the laser, the laser alignment should be double checked and fine tuned via its mounts if necessary so the incident beam is offset 1/4" toward the back of the setup from the center of the CC and aligned with the holes, and thus the return beam will be offset 1/4" toward the front of the setup and parallel to the incident beam.

        • The vertical position of the CC is adjusted by loosening the post holder's thumb-screw and should be set so the incident and return beams are at the same height.

      8. Place a piece of paper where the detector would be for the beam from the turning mirror. Fine tune the orientation and height of the mirror mount so there is a bright return beam there from Arm 2.

  • Arm 1: These steps assemble the components of Arm 1. It may be desirable to block the return beam from Arm 2 until the initial Arm 1 adjustments have been done.

    1. (V2.1 only) Attach a 2" post holder to a BA1S holddown with a 1/4-20 3/8" cap-head screw. Orient it so the thumbscrew (knob) is accessible. Attach this assembly to the baseplate using a 1/4-20 1/2" cap-head-screw.

      (V1.5 only) Attach a 2" post holder to the breadboard using a 1/4-20 setscrew.

    2. KM100 or similar mirror mount: Secure to a 2" post with an 8-32 cap-head screw. Slip the post into the Arm 1 post holder and hand tighten its thumbscrew. Adjust the two alignment knobs so that the mounting plate is parallel to the backplate in both directions. If the adjustments are hard to turn, check that any locking screws are not tightened. Else, total removal, cleaning, and lubricating with a tiny amount of light grease or machine oil will be required.

    3. Cube Corner (CC) trihedral prism: Install a CC in the mount with its apex facing out and oriented so that a flat is at the top or bottom. This is to orient the prism so that the beams do not hit an edge. It should be secured with either a soft-tipped set-screw or Nylon wide-head screw. DO NOT overtighten - it should be snug enough not to fall out (these are fragile!) but not so tight as to smash the CC! Note that the CCs mount backwards from what might be expected so that their edge can be secured securely. :) They thus face through the mount.

    4. Start with the knobs of the adjustable mount set so its fixed and moving plates are approximately parallel and orthogonal to the beam axis as in the diagrams. The BA1S can be moved to adjust the coarse horizontal position of the return beam. The post holder's thumb-screw may be loosened to move the adjustable mount up or down to adjust coarse vertical position of the return beam. The return beam should hit the PBSC at the same height as the incident beam, and offset 1/4" to the left of center. The spacing between the centers of the incident and returns beams should then be 1/2".

    5. Using your piece of paper with the Arm 2 return beam unblocked, there should now be two spots reflected from the turning mirror corresponding to the returns from Arms 1 and 2. (Unless, that is, you're very lucky and they are already superimposed perfectly!)

    6. Adjust the position and orientation of the CCs to superimpose the return beams from Arm 1 and Arm 2. This can be fine tuned with the adjustable mount knobs initially by eye. Then later once the oscilloscope is set up, it can be further optimized by maximizing signal amplitude from the detector.

  • As a quick test, place a linear polarizer (LP or LP/CP with the non-adhesive side facing the beam) with the polarization axis at 45 degrees in the combined return beam. If the alignment is close, very slight rotation of the linear stage micrometer should result in the intensity varying dramatically. Further fine tuning of the alignment may be required to maximize the variation and uniformaty of the fring pattern. The optimal alignment will be where it goes almost totally dark to light. Of course, much more more can be done using these simple observations including effects of alignment, expansion using a lens, display using a Webcam, etc.

  • Set up the oscilloscope: (If using a USB scope, this assumes that the required software and device drivers have already been installed on your PC or MAC.) For input, use one of the scope probes on the 1X setting (there is a slide switch on its body).

  • Single Channel Detector: These steps assemble the components of the fringe detector. Parts for the home-built (custom) detector are normally provided in the kits but a DET110 may be available as an option.

  • Adjust the position of the detector so that the combined return beams are centered on the active area of the sensor.

  • If alignment is close, the amplitude of the signal on the scope should vary dramatically as the micrometer is rotated by the smallest amount - (or from just touching the breadboard or anything on it)! The wavelengths of light are TINY! Each full cycle is 1/2 wavelength or around 316.5 nm. The micrometer moves the stage by 0.5 mm per full rotation, or around 1,389 nm/degree.

  • The amplitude can be maximized using the knobs on the Arm 1 and Arm 2 adjustable mounts. The signal amplitude may vary slightly (up to ~20 percent) in a periodic cycle over a time scale of seconds to minutes in addition to it probably increasing slightly as the laser warms up. The time scale will depend on how long the HeNe has been on. Why? There can be several causes.

    And for those new to interferometers, to reiterate, the optimal alignment will also be where the signal instability is maximized. ;-) Almost ANYTHING will affect it from touching the apparatus or table on which it is on, to just walking across the floor. The wavelength of light is really really small. ;-) To put this in some perspective, a full cycle of the signal with the Linear Interfemeter is a change in PLD of 316.4 µm (1/2 wavelength of 632.8 nm or 1/3,160th of a mm). That's about 1/158th the diameter of an average human hair (~50 µm) or 1/22th the diameter of a human red blood cell (7 µm). Street traffic will be detectable, as will drafts from the A/C, changes in temperature, and siesmic events. Some of these effects can be further explored using parts in these kits.

  • Now you're set to explore all sorts of exciting aspects of interferometry. ;-)



  • Back to Sam's Educational Michelson Homodyne Interferometer Project Manual Table of Contents.

    Observing the Effects of PLD on Fringe Contrast

    The tests above were done with the PLD near 0. What happens otherwise? If the laser operated with a Single Longidudinal Mode (SLM), the PLD would not matter up to a very large number in the 100s of meters or more. (Such lasers are also called "single frequency".) However, the laser used here is NOT SLM but has 1 or 2 modes depending on its cavity length, which changes due to thermal expansion during warmup as the modes sweep through the neon gain curve. (There is much more on this in the section: Linear polarized versus random polarized laser.)

    For the first of the following tests, the lasing modes must all have the same polarization. And the (single pass) Linear Interferometer should be used.

    For all these tests, it will be better to shut off the laser for a few minutes before starting. Then when it is turned on, the mode sweep due to cavity expansion will be fastest.

    1. Check behavior with the PLD set as close to 0 as possible by measurements of the distance from the PBSC to the CCs in each arm. ±1 mm will be acceptable. Monitor the behavior of the detected signal over time by twiddling the micrometer periodically over a few minutes. Note any significant change in signal amplitude. A change of 10 or 15 percent can be attributable to the normal variation in total power during mode sweep and warmup, but anything more will be due to the interferometer.

    The cavity length of the tube in the 1107 and 1108 lasers is around 137.6 mm or 5.417 inches. One half of this is 68.8 mm or 2.7085 inches.

    1. Change the location of the mirror in Arm 2 so the PLD is within ±1 mm of one half the cavity length by relocating the stage on the rail and/or adjusting its position using the micrometer. Now observe the fringe signal again and describe what you see over the course of a few minutes. (Turn off the laser again and allow it to cool for a few minutes as above.) Check alignment to confirm that a change in alignment is not the cause of the effects being seen.

    2. Try intermediate locations for the Arm 2 mirror.

    3. If the Arm 1 adjustable mount is positioned close to the PBS cube using the BA1S holddown, it will be possible to achieve a PLD of at least the tube cavity length. See how the signal amplitude there compares with the one with a PLD of 0. If it's noticeably lower, why might that be? There are several factors involved.

    Now explain the behevior in each case. And what is special about a PLDs of zero and half the cavity length?

    How might these results differ if the HSPMI were used instead of the LI?

    If your laser is random polarized, it is possible to perform the following additional tests with the CP removed from the front of the laser:

    1. Repeat the above tests with the laser oriented so the outer lines denoting the polarization axes of the tube are aligned with the horizontal and vertical axes.

    2. Repeat with the lines on the laser oriented at ±45 degrees. This is probably the most interesting case.

    Explain your results with respect to the longitudinal mode behavior.

    What would happen if the PLD could be extended to more than the cavity length of the laser as would be possible with the 12 inch rail in the Deluxe+ kit?

    All of these tests can also be done with the other interferometer configurations. Predict how the results would change, if at all. What about the HRPMI?



  • Back to Sam's Educational Michelson Homodyne Interferometer Project Manual Table of Contents.

    High Stability Plane Mirror Interferometer (HSPMI)

    The basic Plane Mirror Interferometer (PMI) as its name implies uses a plane mirror instead of a cube corner for the remote reflector. It has a double pass architecture which halves the distance for a full fringe cycle at the detector for a movement of the reflector in Arm 2 (called the "measurement arm"). However, it is not desirable to use a PMI here because it is double-pass only for Arm 2 but single pass for Arm 1. Thus while the PLD can be set to zero, the spacings or lengths of the two arms (as well as the change in displacement) are not the same, which at the very least is confusing. (More on this in the section on the PMI.)

    The HSPMI on the other hand is perfectly symmetric: The beam paths for both Arms 1 and 2 are double pass and go through the CC. However, the change in PLD is double the change in position of the mirror in either arm. Thus it could also be used as a differential HSPMI where the relative displacement of Arms 1 and 2 is to be measured.

    Normally, the Arm 1 mirror would be mounted along with the QWP on the PBSC as the reference since absolute PLD doesn't matter with the single frequency or two frequency lasers used in metrology applications. But as with the LI, we need the PLD to be close to zero or ohter specific value for experiments using a multi-longitudinal mode laser. (The "other specific value" would normally be a small integer multiple of the laser's internal cavity length. Why?)

    As with the LI, above, the designations m-n show the paths taken by the Arm 1 and Arm 2 beams where "m" is the Arm and "n" is the sequence number.

    High Stability Plane Mirror Interferometer Setup

    Assuming the LI was already built, not many changes/additions are required:

    1. Install the HP retro-reflector (10703A) on the PBSC face closest to the detector. The orientation should be such that the beam doesn't hit an apex. This usually means the serial number (if present) runs up and down.

    2. Install QWPs (HP 10722A or unmarked) on the faces toward the Arms. Their orientation does not matter as long as the screw slots are used.

    3. Replace the unmounted cube corner retro-reflectors with circular 1" planar mirrors. The mirror mounts can be rotated 180 degrees to make them easier to adjust.

    Adjust the location of the Arm 2 planar mirror so that the PLD is close to zero. Since the Arm 1 and Arm 2 beam paths are identical, distances from the mirrors to the faces of the PBSC block can be used.

    Alignment will be similar to that for the LI, differ in some respects due to the planar mirrors and double pass architecture:



  • Back to Sam's Educational Michelson Heterodyne Interferometer Project Manual Table of Contents.

    Monitoring the PLD

    The entire purpose of these setups is to measure what happens with changes to the PLD either explicitly or as a result of some external influence. For example twiddling the micrometer stage knob or using the setup as a temperature sensor. Monitoring of the behavior in either qualitatively or quantitatively can be done in several ways. For small changes, direct observation of the REF and MEAS signals is the simplest and often the most precise. But for larger changes, using the µMD2 measurement display will be required.

    Direct Observation

    These use test equipment like the DSO to show what's going on at the very lowest (signal) level. They are most effective for relatively slow changes of up to a few dozen wavelengths, usually found in the various experiments with interferometer sensors. But these can also be the result of turning the micrometer knob or tapping on the breadboard.

    1. Using the DSO to directly view REF and MEAS: This is simplest. Display the waveforms for REF from the laser head cable and MEAS from BPD1 or OR3 above the one-another on the scope. Trigger on REF. Then the MEAS waveform will move back and forth relative to REF as the PLD changes. Almost any type of oscilloscope with a bandwidth of at least a few MHz would be adequate.

      With care, both the gross change of the PLD in integer wavelengths as well as the subtle variation in rate of change at sub-wavelength resolution can be observed either directly or by recording the display and playing it back in slow motion. Since the change is analog - not quantized into bits - more subtle behavior can be easily seen at a resolution down to a few nm. And the extreme sensitivity of the interferometer to any disturbance will be obvious.

    2. REF-MEAS difference: Another method that would provide a display that might be useful is to combine the REF and MEAS signals using a simple resistor network. However, this does lose the direction information. A 10K ohm pot wired with terminals 1 and 3 connected to REF and MEAS, and the wiper connected to a DSO with a 10K ohm resistor to GND would provide a voltage at the split frequency with an amplitude that would vary based on their phase relationship - maximum when in phase and minimum when 180 degrees out of phase. The pot would be used to maximize the variation to account for any difference in amplitude of the REF and MEAS signals. The envelope of the combined signal can then be displayed on the DSO.

    Using a Measurement Display

    µMD0, µMD1, or µMD2 - can provide a numberic readout or plot via the PC GUI or more directly on a DSO with some small additions to the firmware and hardware. OR3 or AB2 will be required to provide the RS422 signals And the resulting display may be more quantized ("steppy") in appearance for small changes in PLD than direct measurement of the analog signals.

    Source code for the GUI can be made available, but changing anything is highly discouraged! Hacking the firmware for µMD0 (Homodyne only) is straightforward. For µMD2 (Homodyne and Heterodyne versions), it is more dicey but shouldn't explode the Universe if care is taken to limit additions to the code just before the values are sent to the USB port - and not messing with any values that impact what is sent! However, there is no - zero - tech support for unauthorized changes, even to a single bit in a comment fields. I've lost too many brain cells in the creation of the GUI and firmware to want to revisit them. And after overloading my brain, many of the relevant memories have been off-loaded to long term storage. ;( :-)



  • Back to Sam's Educational Michelson Heterodyne Interferometer Project Manual Table of Contents.

    Controlling the PLD

    These are various methods of changing the PLD manually and electronically, as well as using the interferometer as a precision sensor. The typical arrangement is that Arm 2 is for the actuators (both manual and electronic) and Arm 1 is for the sensors but that's not cast in stone.

    Before proceeding with any of these, please read the sections on Monitoring the PLD since changing the PLD without measuring it is like a tree falling in a forest with no one around. ;-)

    Thorlabs or Ball Bearing Rail

    All setups after V1.0 come with either a Thorlabs rail (RLA0600 or RLA1200), or a ball bearing rail (MGN15-150mm or MGN15-300mm). The latter is standard with V2.1. The rails allow for coarse positioning of Arm 2 to adjust the PLD. The advantage of the ball bearing rail is smoother movement so that the alignment is not lost and thus the measurement readouts (µMD0 or µMD2) can track without error.

    Micrometer Linear Stage

    All setups come standard with a precision micrometer linear stage that mounts on the rail carrier or carriage. Three types have been used: a Parker 3902 (English) or 3902M (Metric), a generic stage with the knob on the end, and a generic stage with the knob on the side (both Metric). All types have a range of around 1/2 inch or 12.7 mm. The generic stage with the knob on the side is the default type with V2.1. The knob on the side is preferred in part because it won't interfere with the Motion Control Platform's motor housing.

    The micrometer stages permit µm-scale displacement to be manually set, though as a practical matter doing *anything* at that resolution manually is a challenge. And even just touching the knob introduces a displacement shift and vibrations which are easily detectable and jitter in the fringe display or measured displacement.

    Motion Control Platform

    This is currently an option but may be included standard in the future.

    It would replace the Thorlabs or ball bearing rail and carrier with a stepper motor-driven platform and an Arduino-compatible board to drive it. Among other things, this will enable the characterization of non-linearity, backlash and displacement scaling error. These are critical parameters in CNC machine calibration. In addition, controlling Arm 2 motion electronically should introduce far lower vibration than actually touching the micrometer knob.

    However, these are not super precision rigs. Their selection was based on the small size to fit the interferometer setup - and cost. So they may have significant free-play, backlash, and other errors. But that's exactly the sort of thing that experiments can aim to characterize. The main issue is that they use a lead screw (not a ball screw) into a threaded brass block. And aside from a setup like that being inherently imprecise, the machining is sometimes, well, not so great as well. :(

    (Doing something similar using a servo motor rather than stepper motor is possible and theoretically superior, but likely much more expensive and complex.)

    The basic components are:

    The setup below was used for the initial implementation and is similar to what will be included in the kit, (currently an option). This rail with a travel of 200 mm would fit easily fully on the "Extended" version of the interferometer. In fact, rails with a 250 mm travel would also fit and they may be available in the future. For the shorter setups, a rail with a travel of 100 mm would be used. Putting the motor on the left enables a micrometer stage with the knob on the end to be used without hitting the motor housing, which is at a greater height than the platform. However, to minimize the PLD for the homodyne setups, the motor should be on the right using a micrometer stage with side-mounted knob. For Heterodyne, the absolute PLD doesn't make any difference so there's no real need for the reflector to be able to move as close as possible to the interferometer.

    Holes will need to be drilled or drilled and tapped on the breadboard to mate with the Metric hole spacing of the rail. BA-type hold-down clamps or even industrial strength double sticky tape would also work, if a bit clunky.

    To mount the micrometer stage on the stepper platform, an adapter plate is needed unless one is willing to simply use glue. It's the same one used with the ball bearing rail, so that can be transferred if available. 5-Minute Epoxy would be satisfactory, and could be removed if necessary. (If installed on a V2.0 setup, the optical centerline increases by 1/2 inch so some parts change, mostly Thorlabs posts and spacers.)

    The Atmega 328P Nano runs the firmware in the link, below. This is for testing only or if all that's desired is a manually-controlled motorized stage. ;-) It has no error checking so twiddling the knob faster than the motor can move so it just whines is very easy, and there are no limit switches so the flaform can smack into the end plates. But as can be seen, there are many extraneous useless LEDs. ;-) The DM320T controller is set for 400 pulses/mm and the lowest motor current of 0.3 A. Higher current will increase the maximum speed to some extent, but don't go above a sustained current of more than 0.6 A. CAUTION: I you decide to play with a rig like this, make sure the DC to the DM320T is clean and comes on quickly. Else the DM320T may get into a funky state which among other things can send lots of current through the motor, as well as not respond to drive pulses. At least, that's what appeared to have happened on more than one occasion when ramping up the input voltage manually. A 12 VDC wall adapter solved that. I don't know if the red Fault LED on the DM320T being off is sufficient to guarantee correct operation.

                  

    Test of 200 mm Travel MCP (left), on Extended Heterodyne Setup (Left-Center), Test of 100 mm Travel MCP (Right-Center), and on Basic Homodyne Setup (Right)

    The photos show samples of the 100 and 200 mm travel MCPs along with typical installation on the optical breadboards. The electronic Solderless BreadBoard (SBB) with the Nano driving the 200 mm versions is my original rig for testing the quad decoder sketch; the one with the 100 mm versions is all that's needed and is what will be included in the kit. Feel free to add decorative LEDs. ;-) And yes, if you look closely, the tip of the micrometer is not touching the bracket in one of the pics. That's because the stage can be locked in place and wasn't unlocked for the photo. ;-)

    The hardware is the easy part and a mock-up has been constructed as shown above. For V2.1, the motion control platform can be substituted for either the Thorlabs or ball bearing rail with virtually no changes to anything else. But for V2.0, it will not quite be a drop-in as the heights have increased by 1/2 inch. So some opto-mechanical components will need to change - mostly Thorlabs posts and spacers, but there would be negligible additional cost involved.

    The software to make this educational and useful is the challenge.

    An Arduino sketch for an initial version that simply enables the platform to move based on Quad-A-B encoder signals - NOT from the interferometer! - can be found at:

    The SG-µMD0, SG-µMD1, or SG-µMD2 PCBs can have firmware added for motor control in addition to their other duties. Or another microcomputer could be used.

    Suggestions for actual challenging applications (and coding volunteers) are welcome. :( :-)

    Mechanical installation

    1. Motion control platform: If purchased at the same time as the interferometer, holes will already have been drilled in the breadboard for the stepper motor rig, with the motor on the right. These locations will enable the maximum usable travel.

      Secure the platform on the breadboard using four 6-32 1/2" cap-head screws if the holes are tapped, or 6-32 3/4" screws and nuts if not. Take care to assure that it is aligned.

    2. Adapter plate and stage: These are the same as used for the ball bearing rail. If transferring them, first remove the stage and then the flat head M3 screws securing it to the ball bearing rail carriage Reinstall in the opposite order on the MCP using the appropriate set of holes. To maximize range, have the plate extend on the side away from where the knob is located.

    Electrical installation

    Please refer to the connection diagram below.

      Typical Connection Diagram for Motion Control Platform Testing

      (To reverse direction, do only ONE of the following: Swap A and B, Mov DIR from D16 to D14, swap A+ and A-, swap B+ and B-.)

    1. Motor wiring: These may vary. Two schemes are known:

      1. Red/blue, green/black.
      2. Yellow/blue, red/green.

      There could be others, so confirm with a multimeter. The coil resistance should be no more than a few ohms. Connect them to the A+/A- and B+/B- terminals of the DM320T. As a practical matter, it doesn't much matter which motor winding goes to A or B, or even the polarity as long as they aren't cross-connected (which should not damage anything but the motor won't move). Direction of motion is reversed by swapping any single pair of wires going to a specific winding. The DM320T connector blocks may be removed by grasping them and pulling straight ou if that is more convenient to secure the connections.

    2. DM320T power: 12 to 15 VDC is recommended. For the homodyne systems, a 12 VDC power pack is included for this purpose. For the heterodyne system, DM320T power can come from the +15 VDC power supply for the laser.

    3. Control signals: These are opto-isolated logic level signals. OPTO should be connected to +5 VDC from the Nano. DIR(ection) and PUL(se) can be driven from a Nano digital output as in the sample sketch. ENA(ble) defaults to active if left open. PUL triggers the motor to move on the negative edge of the electrical signal (logical rising edge).

    4. DM320T DIP switches: For the NEMA11 stepper motor, set "PK Current" to 0.3 A. If there are problems with vibration, go to 0.5 A but not more. A higher current will simply cause the motor to get hotter even at idle. Initially set the "Pulse/rev" to 400 (which is the coarsest resolution). Rotating the encoder will result in fairly fast movement of the platform. For these settings, all the switches should be ON (down). You can play with the Pulse/rev settings later. ;-)

    5. Arduino wiring: Carefully plug the Nano into the solderless breadboard with its USB connector at one end. Power can come from a USB port or USB charger cube, 5 VDC to 5V, or a 7-12 VDC power supply to VIN and GND. (A USB backup battery pack is not recommended as these may shut down after a few seconds due to insuffient current.) If using VIN, adding a zener diode, three-terminal regulator (e.g., 7808), or some other means to drop the voltage to between 7 and 8 V is recommended to reduce the power dissipation in the Nano's on-board regulator. For example, if using 12 VDC for the DM320T, a 1N4732 (4.7 V) will drop the input voltage to around 7.3 V. 5V from the Nano then powers the encoder and OPTO for the DM320T.

      Here are the required connections. Referring to the sample sketch, there are other pins that may be used to monitor up/down pulses and PWM signals to drive various flavored LEDs for your own amusement. ;-)

        Arduino Pin   Nano Pin   Name   Function
       ----------------------------------------------------------------------------
             D2           5       A     Encoder input
             D3           6       B     Encoder input
            D14          19      DIR-R  Direction reverse to DM320T*
            D15          20      Pul    Motor drive pulse (low) to DM320T
            D16          21      DIR-F  Direction forward to DM320T*
            GND          29      GND    Ground/common for optical encoder+
             5V          27      +5V    +5 VDC for DM320T OPTO and optical encoder
            VIN          30      VIN    +7 to +12 VDC input power (optional)
      

      * Only one of the Direction signals should go to the DM320T. They are complements of one-another. So if the platform moves the wrong way, use the other one.

      + Since the control signals to the DM320T are opto-isolated, the DM320T GND does not need to connect to the Arduino or encoder GND.

    6. Encoder wiring: The typical wire colors/functions are:

         Color   Function
       -------------------------------------
          Red    Vcc (+5 to +24 VDC)
         Black   GND/Common
         Green   Output A (open collector)
         White   Output B (open collector)
      

      Double check with the labeling on the encoder. Tin the wire ends or solder them to #24 AWG solid wire to insert into the SBB. The required pullup resistors are provided by the Nano so no other components are required.

      The encoder can also be used to drive µMD0 or µMD2 with Homodyne firmware. For that, pullups will need to be added to the A and B signals and a reference voltage will be required on the A- and B- inputs at about 1/2 the signal swing.

    When power is applied to the DM320T, the green LED should come on but NOT the red one. If it does, check connections and that power is clean and being applied quickly (not ramping up). The motor should not get more than slightly warm even after being on for a while at the 0.3 or 0.5 A current setting. The motor may make a hissing noise when stationary due to the DM320T's drive. I assume it is some sort of dither to overcome stiction (or something). It is not known yet how much the resulting vibrations will be seen by the interferometer. That may be an interesting thing to quantify. Ears and interferometers are very sensitive. ;-)

    The typical rotary encoder included in the kits has a resolution of 400-600 cycles per 360 degree rotation. So turning the shaft should get the platform moving at a reasonable clip and it is easily possible to exceed the pulse rate where the motor responds. In that case, the motor will stall and whine with the platform remaining stationary. This is a mechanical limitation, not the Nano firmware. The lead screw has a small pitch (1 mm) which is desirable to maximize resolution for interferometer testing. But it also means the motor has to exert itself to produce fast movement. At the DP320T's coarsest Pulse/rev setting, the maximum platform speed is about 1 cm/second corresponding to 6,000 steps/second, or an encoder rotation rate of 2.5 revs/second. (6,000 steps/second / (600 cycles/rev * 4 pulses/cycle). If the DM320T DIP switch Pulse/rev setting is increased, the rate of movement will decrease proportionally and the maximum (input) pulse rate before the motor stalls will increase proportionally. Mounting a massive knob or wheel (not included) on the encoder would enable the friction and imbalance of the shaft to be measured by the interferometer. ;-)

    Voice Coil Linear Actuator

    A loudspeaker consists of a coil of fine wire ("voice coil") on a cardboard form suspended within the field of a strong magnet in such a way that it can only move along its axis. This is attached to the speaker cone. Applying a voltage to the voice coil moves the cone in or out depending on polarity.

    The mini loudspeaker 4 ohm woofer can move its cone a couple of mm with 1.5 V at 0.375 A. But for these tests, it only needs to move a few µm.

    One of the interferometer configurations using planar mirrors must be used for this since it's not practical to attach a cube corner to the loudspeaker. The HSPMI is recommended.

    To use it:

    1. The 1" diameter 1/4" thick "Speaker Mounting Disk" should be glued to the back of the loudspeaker with Epoxy. This will permit it to be installed in a KM100 mount just like a 1" diameter mirror.

    2. One of the small mirrors should be attached to the front of the loudspeaker with the tiniest of tiny dots of Epoxy at the corners. as shown below. Position it to be as parallel as possible to the loudspeaker frame.

    For seeing how its movement will affect the interferometer, the loudspeaker can be driven with a 1.5 V battery (not included) and series resistor. Even with a fairly high resistor value like 10K ohms, the mirror will move decent amount, probably much more than one fringe cycle. Or it can be driven through a resistor via the 10K ohm potentiometer and 10K ohm series resistor from a 9 V battery or 12 VDC power supply. One of these should be included in the kits.

           + o--------+
                      |
         Battery      /     10K
        or Power  10K \<---/\/\---> +
          Supply      /
                      \       Loudspeaker
                      |
           - o--------+-----------> -
    

    Calculate the sensitivity of movement in nm with respect to loudspeaker voice coil current.

    DO NOT connect the loudspeaker directly to the 12 VDC power supply or 9 V battery as they both may be damaged or destroyed. No more than 1.5 V should ever be applied to the speaker. This can be done using a series resistor between it and the power supply or battery or with the potentiometer and series resistor as shown above. Only a very small current will be needed to move the mirror enough to be readily detected and it won't be visible by eye.

    An electronic function generator with simple buffer amplifier and current limiting resistor can also be used to drive the loudspeaker over a larger range as long as the voltage to the loudspeaker doesn't exceed around 1.5 V. With a triangle wave and µMD0, it should be possible to demonstrate the linearity (or non-linearity) of the loudspeaker cone with respect to voltage.

    The loudspeaker will also be sensitive as a microphone so monitoring the detector output on the scope should result in a fairly sensitive response to voice and music, though the frequency response will be terrible due to the large mass of the mirror. Input that to the line input of a stereo system and listen to it on headphones. Why? The quality will be terrible but it will demonstrate possibly the most complex way of going from acoustic sound to electronic sound. ;-)

    Why might any of these NOT behave as expected? Think of not only issues with the interferometer but other causes.

    Piezo Transducer

    A PieZo Transducer (PZT) can be used to move a mirror where only a small displacement is required. (The acronym PZT is actually based on the active material which deforms as a function of voltage, an inorganic compound of Lead (Pb), Zirconium (Zr), Titanium (Ti), and oxygen). Look it up. But our abbreviation is easier to remember. ;-)

    The PZT beeper element in the kit is 27 mm in diameter with an active area of around 20 mm in diameter. It is what's called a "drum head" PZT because the surface moves in and out at its center when a voltage is applied. It can move a few hundred nm with 15 V at essentially no current - it has some capacitance but an infinite resistance, which for slow movement is all that matters. Up to around 100 V can be safely applied to move several µm.

    As with the loudspeaker, one of the interferometer configurations using planar mirrors must be used for this since it's not practical to attach a cube corner to the PZT. The HSPMI is recommended.

    1. A small platform should be fashioned from a washer or something similar so that one of the small mirrors can be glued to the center ONLY of the PZT. This is necessary because a large mounting area may impede its movement.

    2. The PZT can be attached to a 1 inch ID ring (like the lock ring from a Thorlabs 1 inch diameter threaded mount) with 3 or 4 tiny dabs of Epoxy, or can be mounted directly on one of the KM100s in a similar way, or even taped in place. Take care attaching the wires as the solder connections are fragile. Putting a small amount of Epoxy over them (but only in their vicinity so as not to impede movement) is recommended.

    3. A 9 V battery or the 12 VDC power supply can be used along with the 10K ohm potentiometer to vary its voltage:

             + o--------+
                        |
           Battery      /     10K
          or Power  10K \<---/\/\/\---> +
            Supply      /
                        \              PZT
                        |
             - o--------+-------------> -
      

      (The 10K ohm series resistor is not required for the PZT but using it makes the circuit identical and safe for the speaker.)

    Calculate the sensitivity of mirror movement in nm with respect to PZT voltage.

    As with the speaker, an electronic function generator can be used to drive the PZT. With a typical output voltage swing of 20 V peak-peak, the change in position will be several wavelengths. By using a triangle wave and µMD0, it should be possible to observe how good the linearity of movement of the PZT is with respect to voltage, though there will be fewer full cycles compared to the speaker.

    Or it can driven from the loudspeaker output of an audio amplifier through a step-up transformer.

    The PZT may be sensitive enough to act as a microphone as well.

    Gas Cell Compensator

    This may be the niftiest experiment and demonstrates the sensitivity of the interferometer to changes in the index of refraction of air.

    It may be done using any of the interferometer configurations, though the sensitivity will depend on which one is used.

    The concept is that an increase in air pressure will change its index of refraction, and while this is totally invisible to the human eye, the interferometer should be able to easily detect it as a shift in the fringe signal. In fact many fringe cycles even for a space of a couple inches. With some simple calculations, it is possible to corelate the pressure reading on the gauge with the phase change of the fringe signal. If it's sealed well enough, even warming the gas cell by holding it tightly should result in a detectable fringe shift. However, doing that without introducing vibrations that totally swamp any change due to the expansion would be a challenge.

    The Gas Cell Compensator (GCC) consists of a ~2" length of 1" OD Acrylic tube, a pair of planar windows sealed to the ends, the pressure bulb and gauge for a blood pressure cuff (sphygmomanometer), and some simple plumbing. It can mount on a Thorlabs post and post holder using a BA1S hold-down.

    GCC Assembly:

    The Acrylic tube will already be cut to length and drilled and tapped for the 10/32 hose barb and 8-32 set-screw to attach it to a Thorlabs post. The ends will have been ground to be close enough for government work. :) They don't need to be perfectly perpendicular to the tube or parallel to each-other. Nor do they need be polished - the rough cut surface is better for gluing. Only that they can seal to the windows.

    The windows are 1-1/8" in diameter and made of either glass or Acrylic. The Acrylic will have paper protective brown paper on both sides.

    1. If there is more than one hose barb with threads, one of them may have the threads shortened so as not to protrude inside the Acrylic cylinder. Partially thread it into its 10-32 hole. Mix the tiniest amount of 5 minute Epoxy and apply it to the exposed threads. Then rotate it clockwise until fully seated. Wipe off any excess Epoxy and allow time to cure.

    2. Apply a small amount of the remaining Epoxy to the threads of a 8-32 1/2" set-screw. Intstall it in the 8-32 tapped hole so that no more 3/8" is exposed. Apply some more Epoxy at the threads where they meet the tube. (Since this one is vertical and can't interfere with the beams, some portion of it protruding inside the Acrylic tube is OK. Add an 8-32 nut as shown in the diagram for added strength with the Epoxy joining the nut to the cylinder.

    3. Clean one window end one end of the 2" tube with alcohol if available. Soap and water is also acceptable. DO NOT use anything stronger. Make sure it is completely dry and dust-free by shining a bright light through it before proceeding. Repeat as needed.

    4. Carefully place the tube on top of the cleaned window. Put a weight on top of it or use some other means to prevent it from moving accidentally.

    5. Prepare a small amount of 5 minute Epoxy and apply the tiniest bead all around the outside of the joint between the 2" tube and window.

      Avoid getting any Epoxy inside the tube, especially on the windows, as much of their area may need to be unobstructed depending on the type of interferometer and/or whether the laser has a beam expander.

      If it gets messed up before curing, the Epoxy can be careully wiped off and then the glass and/or Acrylic can be cleaned with alcohol (but nothing stronger!). After curing, a single edge razor blade can be used to remove Epoxy, then cleaned with alcohol. Take care to avoid scratching the window(s).

    6. Repeat the previous three steps for the other end. A cotton swab can still be used to clean the previously attached window if necessary. It's especially important that the inner surfaces of both windows be clean as they will be inaccessible once the second window is installed.

    7. Wait an hour or so before proceeding to allow the Epoxy to fully cure.

    8. Connect the hose barb on the tube with the pressure gauge and bulb using the rubber tubing. Test to confirm the connections are reasonably gas-tight and correct if necessary. For this experiment, a small leak may be preferred as it will allow for a controlled pressure decline which can be used to correlate with the fringe signal.

    The photos show the Basic homodyne version with the GCC installed in Arm 1 of the interferometer. The mirror mount post holder is moved further out to make room for it. The arrengement is identical for the heterodyne setups except for the substitution of the laser and optical receiver.

    For more details on the setup to view the signal, refer to the section: Monitoring the PLD.

         

    This photo shows the complete setup with the prototype of the GCC and a scope trace of the photodetector out showing the GCC loosing presure some of its pressure over 20 seconds or so.

    Closeup of GCC Assembly using a Piece of PVC Pipe (left), Overall Setup Showing the Fringe Signal as the Pressure Declines

    Note: If doing this using one of the plane mirror interferometers, the Arm 1 mirror mount may need to face away from the PBSC (with the mirror installed backwards) as with the CCs in the LI to provide enough clearance for the GCC. Then adjust the Arm 2 mirror position so the PLD is 0.

    The blood pressure gauge reads up to 300 mm/Hg (almost 6 psi), but there should be no need to go anywhere near the extreme hypertension region for these tests! :) 100 mm/Hg will be more than enough.

    The gas cell can be mounted in either arm of the interferometer, though using Arm 1 is probably better as it has nothing else. It can be positioned so that either one or both beams (where present) pass through it. (How will this change the calculations?) Avoid aligning the gas cell so that the windows are perfectly perpendicular to the beam paths - angle it slightly so the reflections from the surfaces of the windows do not coincide with the main beams.

    The Arm 1 and Arm 2 path lengths do not need to be the same so that at least is simpler than for the homodyne setup with unstabilized laser.

    Fine tune the alignment of the interferometer to maximize signal amplitude. Close the bleeder valve and slowly pump up the bulb while watching the scope display and pressure gauge. The index of refraction, n, will be approximately equal to 1 + P * k. By measuring the number of cycles and partial cycles as the pressure is changed, it is possible to calculate k. Check it against a value found in a search. Why might it not be the same? Knowing k, an arbitrary pressure can be measured with the interferometer.

    Based on the NIST Refractive Index of Air Calculator using Ciddor Equation, the index of refraction of air at 1 atm (760 mm/Hg), 20°C, and 50%RH, is 1.000271372. As an example, at a pressure above 1 atm of 100 mm/Hg, it is 1.00030715. What is the value of "k"? Over the 3 inches (76.2 mm) inside the GCC, the change in path length is approximately 2.73 µm or 4.31 full wavelengths at 633 nm. You can complete the calculations. ;-) Perform the test with 100 mm/Hg and your favorite interferometer configuration. Explain your results. What are the possible sources of error? Hint: What effect will reflections from the parallel surfaces of the windows have on transmission?

    You might be wondering if it would be possible for the interferometer to act as a microphone using only the change in air pressure from sound waves in one arm. This could be done in principle, but the sensitivity would be extremely poor. In fact to get a detectable response due only to the air pressure variations would require sound levels similar to what might be found a few feet from a jet engine or directly in front of the loudspeaker array at a rock concert. Of course the entire interferometer would be vibrating (assuming it didn't totally disintegrate) and that would dominate any response. Original equipment human ears are extremely sensitive. ;-) See, for example: Engineering Toolbox: Sound Pressure.

    Thermal Expansion

    Or interferometer thermometer. ;-)

    This shows how a change in temperature of an object undetectable by eye can produce a noticeable effect if in one arm of the interferometer. A glass block with two polished surfaces actually called a "compensator plate" is included in the kit, along with 1 or 2 power resistors to heat it.

    It may be done using any of the interferometer configurations, though the sensitivity will depend on which one is used.

    Thermal Assembly Assembly:

    1. Thread the ~1x1-1/4x1/4" block with 8-32 stud into a Thorlabs post and install it in a post holder attached to the breadboard for convenience in mounting the other parts.

    2. Use a drop of Epoxy to secure the 25 ohm 5 watt resistor to the block using one of its flat faces. Wait at least 15 minutes for the Epoxy to cure.

    3. Use a drop of Epoxy to secure a long frosted surface of the compensator plate to the top (labeled side) of the power resistor. Wait at least 15 minutes for the Epoxy to cure.

    4. This assembly can then be installed in Arm 1 of the interferometer using a Thorlabs 1.5" post, post holder, and BA1S Holddown. Angle the broad polished faces of the compensator plate very slightly to avoid back-reflections into the beam paths.

    The 12 VDC power pack is used to do the heating. DO NOT use a 9 V battery, it won't last very long. The power into the resistor(s) is 12*12/R - 5.76 watts for the 25 ohm resistor.

     

    The Coefficient of Thermal Expansion (CTE) for optical glass is around 8x10-6/°C. (It varies slightly depending on the specific type, which is not known for the compensator block.) That means a 1 °C change in temperature will result in its length changing by 8 ppm (parts per million or 0.000008 x its length). Assume that the index of refraction of optical glass, ng, is approximately 1.5. (Again not precisely known .) Calculate the expected number of full cycles from the detector for a 10 °C change in block temperature. Don't forget that it's the net change in PLD that matters.

    Monitor the signal using your choice of method as the block heats or cools and use the results above to estimate the temperature of the glass block based on its length. Without actually knowing the temperature of the block throughout its volume, and knowing it actual CTE and ng. it is not possible to be precise. That's OK.

    The effect will not be as dramatic as with the GCC, above, but with care, should be easily detectable.

    CAUTION: Do NOT leave the resistors plugged into the power pack continuously for too long as bad things may happen.

    What else may be impacting the PLD change besides the block itself? For example, is there any detectable response to the heating if the block is rotated and/or offset so it just misses both beams?

    Index of Refraction of Air

    This one may be a bit more challenging but no additional special parts are required. (And it's related to the last question, above.)

    The index of refraction of air, n, varies by just under -1 ppm/°C. Or more precisely, according to the same NIST Web site, -9.517x10-7. So heated air in one of the interferometer arms should change the path length due to its change in n.

    This can be tested with the same Gas Cell Compensator assembly and interferometer configuration used for the air pressure measurements. Heat it with a blow dryer with the inlet port unplugged from the hose so that the pressure won't be affected. Do this well away from in the interferometer to avoid heating other components. Then quickly install it in Arm 1 and wait a few seconds for the vibrations to die away. Watch the signal as its temperature (and that of the air inside) declines. The heating could also be put into an oven on LOW. Just don't get it so hot that the Epoxy decomposes (which could be as low as 160 °C). :( :) While the sensitivity of n with respect to temperature compared to the effect on the glass block is around 1/10th as great, the GCC is ~5 times as long, so it should still produce an easily detectable signal.

    Note that the expansion of the Acrylic cylinder itself is not a significant factor for these measurements. Why?

    Gas Partial Pressure Measurement (Advanced)

    In addition to temperature and pressure, other gases mixed in with air. or in pure form affect the index of refraction. A variety of common substances have a significantly higher index of refraction than air and thus could result in a detectable effect even at low percentages of partial pressure. These include carbon dioxide, acetone, alcohol, chloroform, and ether. Thus, if one of these is introduced into an arm of the interferometer, there should be a detectable change in PLD. It's probably not a good idea to be messing with chloroform or ether, and even acetone has its risks (both to bodily internal organs and to plastics including acrylic), but certainly CO2 (from carbonated beverages, a CO2 gas canister, or even exhaled breath) and isopropyl (rubbing or medicinal) alcohol or ethyl alcohol (wine, whiskey) can be suitable for tests. :-)

    Engineering Toolbox - Refractive Index for some common Liquids, Solids and Gases lists the values for many common substances.

    Parts to do these tests are not included in the kits, but with a bit of resourcefulness, it should be possible to provide a suitable vessel either for a gas or liquid (with its vapors actually being what's measured). The Gas Cell Compensator, above, can also be used for this purpose if a way is devised to fill it with the test gas, for example by adding a second hose barb so there are entry and exit ports.

    Inexpensive glass cuvettes with polished parallel sides would make suitable containers to introduce liquids, or with an improvised cover, gases, without the need for additional plumbing. Cuvettes are typically 1 cm wide but may be up to 5 cm or more in length. 1 cm is not enough width for both beams in an arm to fit and the length is desirable to maximize the sensitivity. So, two cuvettes side-by-side may be needed. Or a custom cuvette could be constructed from pieces of microscope slides sealed with RTV Silicone.

    As a simple test, start with the cuvette(s) being empty and allow the interferometer to come to thermal equilibrium. Then carefully add some alcohol (at the same temperature and filled to below the level of the beam) and watch the fringe signal as vapors come off the liquid.

    Mounting the cuvette(s) on the power resistor heater could allow the effects of temperature to be explored either with a liquid or gas. But interpreting the results may be more complex than it appears at first.

    More on this is left as an exercise for the student's imagination. ;-)

    Earthquake or Vibration Detector

    If you have successfully constructed any of the interferometers, it will have been obvious that avoiding generating a signal due to vibrations is a challenge even if the entire setup is on a stone countertop. :) But what about actually enhancing this effect? The mirror on the loudspeaker does that to some extent. However, by mounting one of the reflectors (cube corner or mirror) remotely, it will be possible to easily detect someone walking across the room or a truck going down the street a block away. Or an earthquake.

    For these experiments where one of the arms is of considerable length, the 6 mm beam (up to around 5 meters) or 9 mm beam (up to at least 10 meters) will need to be used. A narrower beam would expand too much to be useful. But don't get carried away putting the reflector 25 meters away. Remote alignment will be a large challenge at the very least.

    The best configuration to start with is The LI, but with the genuine HP 10703A CC as the remote reflector and one of the generic CCs attached to the interferometer PBSC for the reference. The 10703A will almost certainly provide better parallelism for the outgoing and return beams than the generic CC. This isn't as important for the short beam paths of the reference.

    Start with perhaps 0.5 meter. Then extend it gradually, tweeking alignment at each step.

    The remote reflector should preferably be mounted on a separate structure, not the same table as the rest of the setup. Another table or wall, for example. Or even suspended by springs or wires. Either arm can be extended but using Arm 2 with the linear stage may be simpler for mounting to a slab of wood or aluminum.

    This rig will be very sensitive regardless of the type of interferometer so anything more complex than the LI may not be justified. Those using a planar mirror as the remote reflector will be more challenging to align because of the mirror rather than CC and also that they are double pass. But if you insist, the PMI should be an acceptable "upgrade" but make sure the 10703A CC is used for the measurement beam. There probably isn't much point to constructing the HSPMI - and alignment will be trickier. Stability is going to be dominated by external factors, but feel free to go there. ;-)

    The use of µMD0 or µMD2 to actually measure displacement would be best as the data could also be captured and (with some simple formatting in Excel) then be formatted like a seismograph. But just watching the signals on the scope should provide some valuable insight into what's going on. Not only vibrations, but temperature changes and even air convection should be detectable.

    Variations on the basic interferometers can be used to sense tilt (angle), lateral movement (straightness), deviation from right angle (squareness), and others. See Optics for Interferometers Using Two-Frequency Lasers. The same optics can be used with single frequency lasers.



  • Back to Sam's Educational Michelson Homodyne Interferometer Project Manual Table of Contents.

    Other Interferometer Configurations

    There are some other variations on the Michelson interferometer that may be put together using parts in this kit, starting with the original form of the Michelson Interferometer.

    Original Michelson Interferometer (BMI)

    This is identical to the original except that it uses the polarized laser and PBS cube as the beam-splitter. But note that the PBS passes and reflects the X (parallel to the baseplate) and Y (orthogonal to the baseplate) polarized components and they retrace their steps on the return journey. Thus a Non-Polarizing Beam-Splitter (NPBS) must be inserted to pick off both of them before passing through the linear polarizer at 45 degrees to combine them for the detector. The NPBS can be the variable Attenuator Plate (AP) that is part of the quadrature decoder, but mounting it and determination of the optimal attenuation is left as an exercise for the student.

    Without the cube corners and/or QWPs to separate the outgoing and return beams, everything will be jumbled together and there will be back-reflections directly to the laser. This won't cause damage but there could be serious instability in the resulting behavior. However, with a polarized HeNe, the effects may not be detectable either visually or even in the fringe signal. But for a random polarized laser, the result would be mode polarization switching, which could reak havoc with the signal.


    No Retro-Reflector Plane Mirror Interferometer (NRRPMI)

    The NRRPMI minimizes the required size of the optical components but with no retro-reflector, will require very precise in alignment during setup to maintain a usable signal with any significant movement. It is most similar to the original Michelson interferometer but the addition of the QWPs avoids (most) back-reflections to the laser.

    Plane Mirror Interferometer (PMI)

    The PMI is probably the most common of the interferometers that use a planar mirror.

    However, it is asymmetric in terms of the beam paths. The reference (Arm 1) is single pass while the measurement (Arm 2) is double pass. To achieve a PLD close to 0 - required for the non-single mode laser - the positions of the two reflectors must differ significantly. Note how close the mirror on the stage is to the PBSC in the diagram - and that may not even be close enough for the paths to be equal!

    For this reason, while the PMI is widely used, the HSPMI is recommended as the one to be built after the LI.

    Single Beam Interferometer (SBI)

    This is commonly used where space is tight since it doesn't require two offset beams. Normally, much smaller PBSC and CCs could and would be used. As with the NRRPMI, above, the use of the QWPs avoids most back-reflections to the laser.

    Modified Linear Interferometer (MLI)

    The MLI adds a pair of QWPs to direct the beam to the detector out the side of the PBSC and is achitecturally similar to the SBI but with offset beam paths that have two additional advantages: (1) retro-reflections back to the laser are reduced further and (2) the beams don't hit the apex or edges of the cube corner trihedral prisms. However, only under very specific cercumstances would the additional cost of the QWPs likely be justified.

    Double Pass Linear Interferometer (DPLI)

    The remote reflector is a cube corner which is better for long distances yet it has double the resolution of the normal LI, the same as the PMI.

    The signal level may be even lower than with the PMI since the CCs are slightly lossier than planar mirrors. A single pass through the silver coated CCs is ~86% resulting in a net transmission of ~74% since there are two passes. For the planar mirrors these values are closer to 90% and 81%, respectively. The CCs also mess slightly with the polarization. With linearly polarized light, the plane of polarization is rotated by ~±10 degrees depending on which set of internal surfaces are involved. What the effect is on the circularly polarized light is not known.

    What would be the effect if only one of the QWPs were present in Arm 1 or Arm 2? Try it!

    High Resolution Plane Mirror Interferometer (HRPMI)

    This doubles the resolution over the PMI or HSPMI. The HRPMI is also high stability because the two beam paths have the same length through the optics, and a PLD of 0 if the distance in Arm 1 and Arm 2 are the matched. This will be much more complex to align and will require a few additional parts and a lot of determination. ;-)

    The HRPMI is essentially an HSPMI in which instead of the return beam going to the detector, it is reflected back into the interferometer, but offset in position by an additional cube corner and traverses all of the optics a second time. So instead of 2 passes, it becomes 4 passes, and the losses will more than double reducing the signal level significantly. In principle, this could be extended to 6 or more passes using a similar approach, but as you will undoubtedly see if you're crazy enough to attempt to implement the HRPMI, it's already tough enough to align.

    Drawing the detailed beam paths for the HRPMI showing how the photons are routed would be more work than it's worth. But since it is equivalent to the HP/Agilent/Keysight 10716A, a Web search will find information, but no need to bother Google, get it at HP/Agilent/Keysight 10716A High Resolution Plane Mirror Interferometer. However, the 10716A is normally used with a two frequency laser for heterodyne interferometry. So, wherever it refers to "ΔF", replace that with "ΔΦ" since we are changing the phase rather than the frequency.

    The HRPMI setup requires some additional optics (another turning mirror and adjustable mount for an unmounted cube corner). The laser may also need to be positioned further to the left to make space. The only way to really test it without a measurement display would be with one of the methods of fine tuning path length - loudspeaker, PZT, air pressure, tmperature, etc. The micrometer stage will simply not have fine enough control to reliably detect a difference between X1 or X4. Thus the setup is shown with the loudspeaker.

    Although drawn with all the beam paths in a plane, it is possible to implement it in 3-D as a 2x2 array within the PBSC by carefully offsetting the cube corners (as is done in the actual 10716A). Consider everything about the HRPMI to be a challenge. :-)



  • Back to Sam's Educational Michelson Homodyne Interferometer Project Manual Table of Contents.

    Displacement Measurement

    One of the most important and wide spread application of interferometry is for precision measurement and positioning. By keeping track of the fringes rather than just observing them, a change in position (called displacement) can be computed down to sub-nanometer resolution. With clever modifications to the Michelson interferometer, angle, straightness, squareness, and others can also be computed. Semiconductor Fabs for every microchip in the Universe utilize "steppers" (a fancy term for multi-axis stages) which position the wafers for exposure are controlled using laser interferometers.

    The output from an interferometer using a single frequency or two frequency laser may be processed to yield displacement information in digital form.

    In their simplest form, the measurement electronics for a homodyne system is just a quadrature decoder circuit driving an up-down counter; for the heterodyne system it is a pair of accumulators and a subtractor. The interferometer optics are identical. In practice, the electronics is considerably more complex, in part to provide sub-wavelength interpolation and extend the range down to nm resolution. And yes, kits are available for these as well.

            

    However, a basic implementation of a homodyne interferometer displacement measurement system for demonstration purposes can be done with a $3 microprocessor board and a few inexpensive parts as shown below. Where the path length difference is limited to be less than a few cm, a multi-longitudinal mode (not strictly single frequency) laser like the one in these kits may be used.

    Quadrature Decodor

    The basic detector using a single photodiode can generate a signal corresponding to light and dark fringes, but cannot provide direction information, essential for using an interferometer in metrology applications. The Quad-Sin-Cos decoder provides a pair of outputs that are 90 degrees offset from each-other in position, similar to the outputs of a rotary or linear encoder:

    This show a rotary optical encoder which uses a pair of LEDs and photodiodes physically offset by 90 degrees to generate Quad-Sin-Cos analog signals which are then thresholded to yield Quad-A-B digital signals. The specific type of sequence is called a "Gray" code (not based on color but attibutable to someone named Frank Gray) and has the property that any possible allowable change in value is a change in only a single bit. This eliminates the ambiguity with sensors using the normal binary order where two bits can change at not quite the same time.

    (The animated encoder graphic seems to be all over the Web. If anyone knows who the original copyright holder is, I will acknowledge them.)

    Many other types of encoders produce similar signals. They may use optical, mechanical, or magnetic sensing, among others.

    An interferometer with angled paths for the two interfering beams produces fringes similar to the pattern of an optical encoder so a quad detector could be built with offset photodiodes. However, more commonly, the 90 degree phase shift is done optically using a single combined beam as shown below. If thresholded and converted to digital form, the result would be a Quad-A-B format.

    This shows variations on one of several common implementations for a Quad-Sin-Cos decoder that provides Sine and Cosine outputs for use in a displacement measuring system. This is among the simplest. In most instances, the photodiodes would be reverse biased to provide a linear response. It may be possible to get away without that for initial testing but it will probably be needed if doing anything useful with the outputs. In addition, a third "Intensity" channel is almost always included to accommodate variations in detected power due to the laser aging, changes in alignment, and contamination over time. The Intensity channel can be implemented electronically or optically with a non-polarizing beam-splitter at the input and additional photodiode.

    The output signals from these will be close to sinusoidal with a relative phase close to plus or minus 90 degrees depending on the. direction of motion of the remote reflector or ring laser gyro.

    The purpose of the angled arrangement is to minimize the difference between the amplitudes of the two polarizations. Otherwise, with 45 degrees being close to the Brewster angle (around 57 degrees), one will be much larger than the other. Even so-called 50:50 beam-splitters may be subject to this, so using the angled arrangement for either one may be beneficial. The parts for the version using the Attenuator Plate (AP) are what are in the kit, which simplifies construction. Using the AP also permits the relative amplitudes of the Channel A and B signals to be changed somewhat without electronic adjustments.

    Some resourcefulness will be required to mount the parts in this kit to put together a Quad-Sin-Cos decoder. A variable attenuator plate is included that may be used as the NPBS. Pieces of CP will be satisfactory for both the combination of the LP+QWP (since that's exactly what the CP is), as well as the LP (flipped) since the output polarization doesn't affect PD behavior. See the information on polarization, below.

    This would be a great excuse to finally make good use of that 3-D printer sitting idle. ;-) A simple frame could be designed to mount the AP via its spring and screw so its position, and thus reflection and transmission, would be adjustable in the beam. Slots and/or faces would be used to attach the pieces of CP and the PDs. Be creative! This is probably overkill though.

            
    Quadrature Detector using Variable Attenuator Plate and Circular Polarizer Sheets. Optical Layout, Parts), Constructed on Solderless Breadboard, Protoboard, and PCB Assembly

    These photos show a diagram for the preferred implementation of the Quad decoder itself, the typical parts, and 3 perfectly workable construction options, the first of which uses a small solderless breadboard and doesn't require any soldering. For that one, the AP and pieces of CP sheet could be glued to wires that would be stuck in holes. Or U-shaped pieces of wire could simply be stuck in holes to keep the CPs and AP in place. ;-)

    Before constructing the prototype or PCB versions, it was dedirable to to conclusively prove that the simple Type 3 scheme with CPs for the polarization optics actually worked as advertised, so a prototype version was installed on the Michelson Interferometer test-bed:

    In the interest of expediency, it cheats and used an NPBS rather than a plate beam-splitter or variable attenuator, two Thorlabs DET110s rather than bare photodiodes, but the QWP+LP for Channel B is a piece of CP (as in the diagram on the right, above) stuck to a microscope cover slip that is glued to a platter clamping ring from an ancient defunct harddrive. Got that? :)

    The ugly scope screen shots below were taken using this setup:

           

    Displacement Positive (Left) and Negative (Middle), and Lissajoous Display

    For the sinusoids, capturing a decent photo while twiddling the micrometer screw is quite challenging. ;-) But the conclusions are clear: This simple Quad decoder does its job well with a phase shift of ±90 degrees. If the Arm 2 mirror or retro-reflector were on an electronically controlled positioner like a loudspeaker voice coil or linear motor driven with a ramp, the waveforms would be textbook quality. ;-) But with only a small stretch of the imagination, it can be seen that the screenshots agree with the expected behavior based on the diagrams, above.

    The prototype on a prototyping board was then constructed and tested to confirm similar behavior, and then the simple PCB was made so that mounting of the photodiodes and other electrical components would be simplified. There is nothing to really secure the AP/BS but that could be done with double-sided tape.

    The Lissajous display on the right was done at a later time under more controlled conditions using a loudspeaker voice coil driven by a triangle wave from a function generator! Using a High Stability Plane Mirror Interferometer, each 360 degree rotation represents a displacement change of 158 nm.

    It has been suggested half jokingly (by me!) that a Rube Goldberg measurement display could be implemented by pointing a video camera at the scope screen and analyzing the image frames in real-time. At least for slow displacement changes!

    Constructing the Quad-Sin-Cos Detector

    The mounting scheme doesn't need to be fancy or pretty but should hold the pieces securely while maintaining alignment. This can use bits of tape and Epoxy or other adhesive. The CPs, QWPs, and NPBS plate are expendible so feel free to chop them up if necessary for them to fit. :)

    The photos above show various possibilities not involving a 3-D printer :), including a simple PCB (which is available), but some soldering is required for that. The simplest approach is to use the same Solderless Breadboard (SB) as the Single Channel Detector, attached to the post using the Detector Adapter Plate as in this closeup:

    1. Set up the HeNe laser so that there is a linearly polarized beam. Orient the polarization to be at 45 degrees or preferably, the output of the interferometer as well.

    2. Mount the AP at an angle as close to normal to the incident beam as is convenient to still be able to locate the Channel B CP and PD in the reflected beam. The rational for this is to minimize the effect of the angled plate on the polarization of the reflected and transmitted beams. Why? If you're into fancy 3-D-printed mounts, it can be arranged to use the screw/spring combination to adjust it's position, and thus attenuation (or transmission and reflection). But that certainly is not required and the screw and spring can be relegated to your junk drawer. ;-)

    3. Mount one PD to intercept the transmitted beam. (This will be designated Channel A or 1.) Trim the leads if necessary so that in conjunction with adjusting the height of the platform, the beam can be approximately centered vertically in the AP.

    4. Adjust the position of the AP so that the transmitted and reflected beam intensities are approximately equal. This will be near or at the higher-density end of the AP. Perfect balance may not be possible since the maximum reflectance of the AP is not 50 percent, so the remaining adjustment will need to be done with the Gain trim-pots (substituting for the fixed resistors, or later when the interface circuit is built).

    5. Install the second PD (Channel B or 2) to intercept the reflected beam. Note that the AP glass is slightly tilted so it may be necessary to trim the PD's leads or bend them over so its height will be correct.

    6. The PDs should be reverse biased with 5 to 15 VDC (+ to cathode, the right leg facing front). A load resistor of approximately 100K ohms should result in a decent signal amplitude for a ~1 mW laser. If too high or too low using only the laser, no need to change until using the actual interferometer output. To accomodate lower power lasers and/or losses in the interferometer optics, 1M ohm trim-pots are included in the kits for this purpose. The typical wiring is shown below.

             R-Protect    PD1      Yellow
         +-----/\/\----+--|<|---<<-----------------------+----------o Scope Channel 1
         |             |                                 |
         |             |  PD2      Blue                  |
         |             +--|<|---<<---------------------------+------o Scope Channel 2
         |                                               |   |
         |                                               /   /
         |                                       R-Load1 \   \ R-Load2
         |                                               /   /
         |                                DC Power       \   \
         |                          Red    +| |-   Black |   |
         +-----------------------<<---------||||---------+---+-------o Scope Ground
                                            | |
      
         |<--- SBB or QDx PCB --->|<----- Scope / Power Wiring ------>
        

    PD Pins: Facing Front of PD with Legs Down: Anode on left, Cathode on right.

    R-Protect: 250-1K ohms typical. R-Load: 10K-1M ohms typical.

    In the photo, above, one piece of CP sheet is stuck directly to the channel 1. PD. The other piece of CP sheet is simply propped in front of the channel 2 PD with the sticky side facing out. Long term, that side should be protected with some 5 Minute Epoxy or a microscope cover slip. The AP is just sitting on the SB. Dabs of 5 Minute Epoxy, wire loops, or other means can be used to secure them more permanently. If using wire loops, take care not to short out anything that shouldn't be connected. ;)

    The following must be done using the output of the interferometer that has been properly aligned so that interference can be seen on a white screen if a linear polarizer is placed in the output beam at 45 degrees.

    Make sure you hands are clean or use a pair of latex surgical gloves when handling the pieces of CP.

    1. Cut a pair of 0.5x0.5cm or larger pieces of the CP sheet aligned with the original edges.

    2. Carefully remove the protective film from both sides of the cut pieces. One side is simply bare plastic; the other side has a sticky adhesive.

    3. The adhesive-side of the CP is the QWP with its optical axis at 0/90 degrees (aligned) with the edges; the other side is the LP with its axis at 45 degrees. This can be confirmed by testing with a linearly polarized laser or a random polarized laser with a separate linear polarizer in front of it:

      • When the CP's LP-side faces the laser, rotating it will result in the transmitted intensity going very close to 0.

      • With the CP-side facing the laser, there will be little or no change in intensity.

    4. Channel A: Attach the adhesive side of one of the pieces of CP directly to the front of the Channel A PD making sure to keep it aligned with the horizontal and vertical axes. This will result in its LP-side facing the AP oriented so the polarization axis is at 45 degrees (edges aligned with the X and Y axes). The result should be a strong signal as the micrometer stage is moved or whatever is convenient to change the PLD.

    5. Channel B: Attach the other piece of CP in the reflected beam so that its QWP-side faces the AP and is oriented at 0/90 degrees (edges also aligned with the X and Y axes). This is best done by using a drop of 5 Minute Epoxy to stick it directly to the Channel B PD, but if large enough, it can just be placed in front of the PD with a wire loop or something similar to stabilize it. Just make sure that the wire loop doesn't short out something important. ;) The result should be a strong signal as the micrometer stage is moved or whatever is convenient to change the PLD. Once correct behavior is confirmed, coat the adhesive-side with a thin layer of 5 Minute Epoxy to prevent it from attracting contamination like dirt and dust, or stick it to a microscope cover slip.

    6. Fine tune interferometer alignment to maximize signal amplitude and adjust the values of the PD load resistors so that the signal amplitudes are between 3 and 5 V p-p.

    7. Observe the Channel A and B signals on the oscillscope. They should be in quadrature (or very close). This means the phase shift between them should flip from +90 to -90 degrees depending on the direction of motion as in the photos, above.

      Note: Even if the peak-peak amplitudes are made equal, it may not be possible to avoid an offset on one channel. In that case, the scope vertical position can be set to superimpose them on the screen, and later, the Threshold trim-pots can be set appropriately.

      There can be a number of reasons for this offest. What might they be? Hint: Think polarization.

    Micro Measurement Display 0 (µMD0)

    µMD0 is a very inexpensive system for precision readout of displacement (change in position), angle, straightness, and more in metrology applications using (homodyne) interferometry with single frequency HeNe lasers. Where the path length difference is limited to a few cm, it may also be used with (unstabilized) multi-longitudinal mode HeNe lasers. µMD0 may be used with devices like linear and rotary encoders which produce Quad-Sin-Cos or Quad-A/B signals.

    For the purposes of these interferometer kits, µMD0 consists of three parts:

    1. Interface from Quad-Sin-Cos analog signals to Quad-A-B digital signals, consisting of very simple circuitry to provide gain and thresholding.
    2. Atmega 328 Nano 3.0 (or similar Arduino-compatible) microcomputer board, which runs the µMD0 firmware.
    3. Micro Measurement Display (µMD) Graphical User Interface which typically runs on a Windows PC or laptop.

    The general organization of a typical system is shown below (though the one implemented in this kit differs in some subtle details):


    Typical Homodyne Interferometer Measurement Setup using µMD0

    To convert the analog sin and cos signals to something for a low cost microprocessor with adequate performance requires a simple interface which provides gain and threshold adjustments. (While it has analog inputs, their conversion rate is way too slow.)

    Referring to the schematic, the trim-pots on the left are the load resistors for the quadrature detector. The 100K value should be satisfactory to resultin a signal of a few volts p-p using a laser with an output power of around 1 mW and no beam expander. (The expanded beam may slightly exceeds the dimensions of the photodiodes so the sensitivity will be reduced.) For a lower power laser or a laser with a beam expander, larger values may be required. Or for finer control, fixed resistors can be added in series with the trim-pots. The trim-pots on the right adjust the comparator threshold for the Sin and Cos signals from the Quad decoder, with the feedback resistors providing some hysteresis. The Atmega 328P Nano 3.0 board runs firmware that is compatible with the µMD GUI. Of course, no high tech system would be complete without indicator lights, so LEDs are added to monitor the A and B inputs. ;-)

    The minimal implementation is shown below along with a shot of the laptop screen while twiddling the linear stage micrometer:

       

    µMD0 and Interface Schematic (left), Impementation on Solderless Breadboard (Middle), µMD GUI Display while changing Displacemet (right)

    A schematic with slightly more detail like pin numbers may be found at: µMD0 Sin-Cos Analog and RS422 Digital Interfaces.

    A simple assembled PCB for quad decoder is included. Blank PCBs are available for the µMD0 and interface, but should not be required for a student project. ;-)

    The ptototype setup is shown below, but with the dual Thorlabs DET110 detectors and interface without gain adjustments, sorry. ;-)

    Overall Setup showing Interferometer, Scope, µMD0 with Interface, and µMD GUI Display

    The bandwidth of the photodiode + resistor combination is quite limited, probably a few thousand counts/second, if that. But it is sufficient to track the movement of the micrometer stage, though not if it's pushed back and forth by hand without the micrometer or being moved on the rail. That would require a proper transimpedance pre-amp circuit. µMD0 has a maximum slew rate believed to be above 125,000 counts/second, or around 1 cm/s with the Linear Interferometer.

    For more information, see the Laser FAQ chapter Laser Instruments and Applications, sections starting with "Interferometers for Precision Measurement in Metrology Applications". And the Micro Measurement Display 0 (µMD0) Installation and Operation Manual.



  • Back to Sam's Educational Michelson Homodyne Interferometer Project Manual Table of Contents.

    Miscellaneous

    Lower Cost Interferometers

    One of the major expenses in assembling these kits is the interferometer itself including the HP/Agilent/Keysight 10702A or 10706A PBS cubes, 10703A CCs, and optional 10722A QWPs (the latter also known as "Plane Mirror Converters"). The typical price for complete HP PMIs is $200-$300 or more on eBay. Even that isn't too bad as purchasing the individual optical parts (1" PBS cube, 1" CCs, 1/2" or 1" QWPs) from places like Thorlabs or Newport would be even higher. (To satisfy your curiosity, the price for a complete PMI from Keysight is over $4,000!)

    Here are two possibilities:

    For interferometers requiring QWPs like the PMI and HSPMI, HP 10722As may show up on eBay at reasonable prices. As an alternative, I have a limited number of 1/2" optical grade mica QwPs which have been tested to work well. These are very fragile so they would need to be attached directly to the PBSs with index-matching cement (5 minute Epoxy may also be suitable), to a plate with holes for the beams, or to plastic or metal washers. Two of these would be required to substitute for each 10722A.

    As partial confirmation, a test was done replacing the CC and 10722As of the HSPMI (but using a 10702A PBS since I don't yet have a generic 1" PBBS). The CC and QWPs were glued to HP alignment plates that attach to the PBSC with magnetic strips as can be seen below:

    HSPMI using 10702A PBSC with a Generic CC and Generic QWPs

    It does work. However, as expected, the signal amplitude is down by more than 65%. This is probably mostly due to the 16 traversals of the uncoated surfaces of the QWPs, followed by 4 traversals of the uncoated surfaces of the CC and 6 reflections from their internal silver-coated surfaces. But there are likely also contributions from not quite accurate 45 degree QWP orientation and less than perfect QWP 90 degree retardation. But the QWPs are mostly responsible for the decrease in signal level, so AR-coated QWPs would help a lot. It's *only* 8 passes but through 16 surfaces.

    The HSPMI is the most complex of the common interferometers; The LI is trivial in comparison with only a single CC. While tests have not been performed with a generic CC attached to the PBSC in an LI, it is expected that the only effect would be a reduction in signal amplitude of around 9 percent due to the slightly lower efficiency of the generic CC of around 0.86 compared to 0.90 for each pass compared to one from HP.

    Next up: Using a non-HP PBSC, though I do not expect there to be any significant difference in performance, at least for the small PLDs of this setup. Stay tuned.

    Where Does the Missing Light Go?

    This question inevitably comes up with interferometers where visible fringes are produced. Where does the light go that would be in the dark spaces? In principle, with perfect alignment, a visible display could be totally bright or totally dark. It is possible to demonstrate this here by viewing the fringe display directly on a screen behind a linear polarizer oriented at 45 degrees. And by rotating the LP between +45 degrees and -45 degrees, it would go between light and dark.

    The analog with the signal is that it can be high or low. But since the laser is still lasing and with no polarizer in front of the detector, the detected signal is more or less constant. So when the signal is low (or the spot is dark), where is the missing power? I'll save you some of your brain cycles and state that with the LP film, it's lost in the plastic and actually increases its temperature a miniscule amount.

    But what about a Polarizing Beam-Splitter (PBS) like the large one that is the heart of this setup? They have negligible losses and it would be simple in principle to use one in place of the LP film. To do that, either the PBSC would need to be rotated 45 degrees or the polarization of the beam would need to be rotated 45 degrees so that the PBS can generate the two polarized outputs. Rotating the polarization of the beam can be done with a Half WavePlate, but there is none in the kit, believe it or not. ;-) However, it can be simulated using two pieces of LP film and the Attenuator Plate (AP) as a Non-Polarizing Beam-Splitter (NPBS). This isn't quite identical because there will still be losses in the LPs, but if oriented at ±45 degrees, the effect will be the same.


    Dual Polarization Detector using LPs at ±45 Degrees

    This uses the same CP as the quad decoder but for these experiments, the QWP portion is totally irrelevant and the QWP (sticky) side goes toward the photodiodes for both.

    It should be pretty obvious what is going to happen, but seeing it is not quite the same as theory. ;-)



  • Back to Sam's Educational Michelson Homodyne Interferometer Project Manual Table of Contents.

    More Advanced Options

    Some, many, or most parts for these are NOT included in the standard kits by may be available.

    Other (non-Displacement) Interferometers

    While a number of different interferometer configurations have been described above - and you may have constructed more than one - they are all used to measure displacement - and its derivative - linear velocity. With the exception of the High Resolution Plane Mirror Interferometer (HRPMI), all use the same basic layout with the PBSC, zero to three optics attached to it, and the cube corners or plane mirror reflectors. The HRPMI requires an additinal cube corner and turning mirror but is otherwise similar.

    But many other parameters can be measured using interferometry including angle, straightness, and flatness. The layouts for these are not particularly complex but would require creativity in mounting and some optics and mechanical parts not included in the standard kits. See the information at Optics for Interferometers Using Two-Frequency Lasers.

    DIY Single Frequency (SF) HeNe Laser

    By replacing the common HeNe laser head with a stabilized single frequency laser, PLDs of up to 10s to 100s of meters would be supported with essentially no other changes. They would become very boring. :) But that's exactly what is needed for most metrology applications. (Strictly speaking, no laser is truly single frequency but rather Single Longitudinal Mode or SLM, which is pretty darn close to SF for all practical purposes. But whatever it's called, HeNes are close to ideal in this respect.)

    With an SF laser and the Quadrature Detector, the signal output from any of these interferometer configurations will provide complete displacement information that can be used with a measurement display or for closed-loop control. Systems using SF lasers are called homodyne interferometers.

    SF HeNe lasers are available for order of $5,000 from a few laser companies (though this number has been dwindling). But fortunately, it is possible to construct one from readily available parts for less than 1/20th as much. The laser tube can be identical to the one used in the JDSU 1107 or 1108 random polarized HeNe laser head that comes with some of these interferometer kits. Adding a heater to control cavity length along with a simple controller using discrete analog components or an Arduino turns it into an SLM laser with performance similar to that of the high priced ones. The laser tube could mount on the same 4-screw rings as the normal head cylinder, but inside its own cylinder along with a thin film heater, similar to what is shown in the photo below under "Future Options" and in the "Setting the Heights" section above. Some basic optics and a small PCB for stabilization would then be added. It's really that simple!

    If interested, a kit of parts along with detailed assembly instructions is available. For the manuals, go to Sam's Electronics and Laser Kit Information and Manuals.

    Modifying the Output of a Two Frequency Laser for use in a Homodyne Interferometer

    In order to utilize the HP/Agilent 5517 laser for experiments in homodyne interferometry, its output must be converted to single frequency. Even for the homodyne kits, this may be desirable if you just happen to have a working HP 5517 or similar laser available would like to ditch the PLD constraints. ;-)

    There are a couple of ways of doing this without actually rebuilding the laser itself. ;-) But unfortunately, both lose at least 1/2 the optical power. There's no easy way around that for a single axis as is the case here. (For two axes, a PBS could be used to separate the H and V polarized components with each one rotated to the desired polarization orientation or converted to circular polarization with waveplates. Each axis could then use 1/2 the total power and nothing would be inherently wasted.)

    1. Linear polarizer at 45 degrees: This is simplest but strictly doesn't create a single frequency output. In fact, the output is similar to that of a multi-longitudinal mode laser like those used in the other kits, but the mode spacing is at the split frequency rather than close to the ~1 GHz of a short linearly polarized HeNe. If the detector is "blind" at or above the 1.5 to 2.0 MHz split frequency of the two frequency laser, this scheme will work fine. But that cannot be guaranteed even for the basic detection schemes used here. And the soon-to-be-released AB2 Quad-A-B preamp has a response exceeding 3 MHz. In fact, as should be recalled, an LP at 45 degrees is exactly the scheme used to combine the F1 and F2 components for the optical receiver. The only reason this isn't an issue with the other kits is that none of the electronics responds at ~1 GHz.

    2. Linear polarizer at 0 or 90 degrees with Half WavePlate (HWP) to orient the beam at 45 degrees: This one is "pure". It blocks F1 or F2 (which one doesn't much matter) and then rotates the single frequency beam to 45 degrees.

    In the interest of efficiency (since as noted, the theoretical best that can be done is to only cut the power in half), a small PBS cube is used rather than LP sheet. The HWP has only a 5-10 percent loss and will be attached to its output-side. (Or vice-versa as the operations are symmetric). The detailed mechanics are yet to be worked out but any special parts will be provided with the combined kit and the "Single Frequency Converter" will mount on the faceplate of the laser.

    DIY Two Frequency Zeeman HeNe Frequency Laser

    This diagram shows the most common implementation of a two frequency HeNe laser, which is based on Zeeman splitting.

    Where a random polarized HeNe laser tube meets certain requirements, applying an axial magnetic field will result in the normal single longitudinal mode splitting into two components that are left and right circular polarized, which are converted to orthogonal linear polarization with a QWP. It turns out that many of the HeNe laser tubes that used to be used in 100s of thousands of supermarket checkout barcode scanners satisfy these requirements, and the laser tube found in a $10,000 metrology laser is essentially the same. ;-) Stabilization is then similar to that of SF laser, above, and a kit is also available. Since the SF and TF lasers are very similar, it may be possible to use the same tube and controller. But while a homodyne interferometer can work using an unstabilized laser with restrictions on PLD, a two frequency Zeeman HeNe must be stabilized because the beat frequency actually appears for only a portion of mode sweep. So it must be locked to the center of that range.

    Kits are available to construct stabilized Zeeman HeNe lasers that would perform well for use with the interferometer. They include a bare HeNe laser tube, power supply, heater, magnets, optional parts to construct an Arduino-based controller, and more. So if you really want to be able to claim that this was built from parts closer to stone tools and bear skins than HP/Agilent lasers, see the manual at: Stabilized Zeeman HeNe Laser Kit 1 with or without Arduino.



  • Back to Sam's Educational Michelson Homodyne Interferometer Project Manual Table of Contents.

    Future Options

    Transparent Laser Enclosure with Heater

    Modified laser head cylinders may be available made of transparent aluminum (a.k.a., acrylic or Plexiglas). While there isn't a huge amount of action that would be visible, it would add a cheery glow to the setup. And at the same time (and more importantly), a thin-film heater can be added to the laser tube to control cavity length. The photo on the left is a closeup of the laser while the one on the right is the rail setup with Thermal Expandsion experiment using the glowing laser.

     

    HeNe Laser Tube in Transparent Acrylic Cylinder with Heater (left), Scope Display of Fringe Signal while Heating Glass Block in Arm 1 (right)

    Something like this would be used to house the single frequency or two frequency lasers described above. The heater would also be useful to accelerate mode sweep for experiments dealing with the interferometer response particularly for random polarized lasers.


    Information

    Required Tools, Equipment, and Supplies

    Not everything here will be needed for all setups but there are the type of items that should be in a well equipped toolbox or on the workbench.

    Parts Identification

    Captioned photos of most of the types of parts used in both the Homodyne and Heterodyne kits may be found at Michelson Interferometer Kit Parts Identification.

    Polarization Control Optics

    Three types of optics that affect the polarization are used in various parts of the interferometer:

    Some specific applications:

    And as a point of interest CPs are used in front of LCD displays to increase their contrast under ambient illumination: With the LP side toward the viewer, ambient light will pass through it and the QWP, and be reflected by the display itself. But when CP light is reflected, the "handedness" flips and the result is linear polarization at 90 degrees to what it was originally, and thus blocked by the LP. That is part of the reason your smart phone screen looks dark when nothing is displayed on it.

    However, while circular polarizers intended for use with cameras are constructed as shown above, the "polarizing films" sold for smart phones and the like may or may not be true CPs. Some are just linear polarizers or linear polarizers coated with a weak neutral density filter. And it's not clear how one can select the proper type since the sellers are typically clueless. The only actual CPs consistent with the above diagram I've found so far were listed as something like: "New Backlit Screen Modify Part Polarizing film For GBA GBC GBA SP N_ES". Those listed as "LCD Polarizer Film Polarization film Polarized Light Film For ip.ccHHH" may also be CPs but with (1) the adhesive and non-adhesive sides swapped and (2) the axes of the LP and QWP rotated 45 degrees. They may also have a weak ND filter. Sometime a faint colored line can be seen which is at the orientation of the LP or 90 degrees from it. But that doesn't help with whether there is a QWP, though sometimes there are also a series of small circles, presumably to indicate CP, though they tend to be on the opposite side from the QWP. Those listed specifically for iPhones were of the simple LP type (usually with a faint line at 90 degrees to the LP axis parallel to the short side). Interestingly though, the protective film that normally gets discarded is birefringent and may even be a QWP. Which of course makes little sense. :( :) Confused yet? I sure am. :( :)

    This same scheme is also used in what is sometimes called a "poor man's optical isolator", whose purpose is to minimize back-reflections from an optical setup into the laser, which may destabilize it or worse. The combination of a linear polarizer (or PBSC) and QWP acts as a "diode" for polarized light. It's called "poor man's" because it is much less expensive than a Faraday isolator, and adequate for many purposes. But for it to work well, any reflective surfaces in the optical setup must not mess with the polarizaton.

    The piece of CP provided in these kits is intended for such an application. It can be cut into smaller pieces since they only need to be slightly larger than the beam, which even if expanded is only ~4 mm. As noted shown in the diagram above, the optical axes of the QWP are at 0 degrees with respect to the edges of the CP and the axis of the LP is at 45 degrees.

    The CP comes with protective film on both sides which MUST be removed because it acts like some type of waveplate and messes with the polarization. The side of the CP with the QWP is sticky since that was intended to go against the screen. To prevent it from collecting dirt and fingerprints, the smaller pieces to be actually used in the detector(s) should be stuck to microscope cover slips (included) or glass windows. CAUTION: Microscope cover slips are thin and fragile - don't press too hard.

    Adjusting the Micrometer Stage

    The Generic or Parker linear stages should move silky-smooth both with the micrometer and if simply pushed up and back. But if there is any roughness - or if the fit appears to be loose, the bearings may be fine tuned without disassembly via recessed setscrews on the side of the moving part of the stage.

    A 1/16" hex driver may be used to tighten or loosen them the smallest amount which should not be more than 1/10th of a turn.

    The micrometer can be lubricated with light grease if it seems rough or tight. The ball bearings for the stage itself should not need lubrication unless serious dust or other contamination has gotten into the tracks.

    Adjusting the Motion Control Platform

    Since the two types of MCPs used here are not exactly super high precision $20,000 parts :), they may suffer from a couple of maladies that can usually be largely corrected with minimal effort. One of these is detectable axial free play, which means that there will be a dead zone if an attempt is made to move the platform by a small amount. Free play is easily detected - the platform itself will move by a small amount if gently pushed back and forth. This may occur after awhile or even when new due to less than stellar quality control. Another issue is uneven torque required to rotate the screw to move the platform depending on where the platform is on the lead screw. It is usually possible to largely remedy both of these with small adjustments of the end-plates and the angle of the platform on the moving carriage. The following applies to both the small and large MCPs, but excessive axial free play is more likely to be an issue with the small ones:

    1. Move the platform to the motor-end of the rail by rotating the lead screw manually or with power applied temporarily to the DM320T controller using the Arduino MCP test sketch and encoder. If the lead screw appears to bind there and is more difficult to rotate or the platform gets stuck, loosen the two screws securing the motor and adjust the lateral position of the motor so that the rotation requires the least torque.

    2. Move the platform to the far end of the rail in a similar manner and repeat by adjusting the lateral position of the bearing end-plate.

    3. Move the platform to approximately the center of travel.

    4. Loosen the 4 screws securing the platform to the carriage and then set them so they are just snug.

    5. Gently twist the platform while testing the axial free play and set it so the free play is just barely detectable. Tighten the 4 screws and confirm that it hasn't changed. If the free play is set to exactly zero, then the the lead screw may be too tight. This is basically compensating for imprecise machine work in cutting the threads.

    6. Run the platform fron end-to-end using a low current setting of the DM320T (e.g., 0.3 or 0.5 A) to confirm it doesn't bind anywhere.

    7. Repeat all these steps as necessary to optimize.

    It still won't be a $20,000 rig but should be quite acceptable.

    Removing 5 Minute Epoxy

    So you took great pains to attach the turning mirror but then realized it's centered on the bracket instead of being closer to the rear. (It may still work OK where it is so there may be no need to panic just yet!). But just in case:

    Where the adhesive is accessible, a single-edge razor blade can often be used to slice it or get underneath and peel it off. But that may not be possible if the turning mirror were attached in the wrong location with bits of Epoxy underneath it. Or reusing the mirror stuck to the PZT. Attempting to pry the thin mirrors off would likely result in bits of glass.

    The easiest method where the parts can withstand it is to use a common heat gun. The exact temperature at which the typical Devcon 5 Minute Epoxy decomposes is not known. The datasheet only states that the "Operating Temperature" range is -40 °F to +200 °F but not what happens at +201 °F. ;-) It may exceed +500 °F. However, from experience, a sufficient temperature can be reached in a couple minutes to soften the Epoxy without damaging parts made of glass or metal. However, some plastics might melt. ;( As the critical temperature is reached, the Epoxy will soften, so work over a non-flammable surface that won't damage the part to be removed and gently prod it with a popsicle stick or something similar until it falls off. After cooling, the residue can be removed by scraping or with isopropyl alcohol. Since the Epoxy at least partially decomposes, this method should not be used for repositioning, only removal. Clean the two surfaces and then start fresh. ;)

    CAUTION: Don't try this with the gas cell or loudspeaker, they would likely get damaged.

    Linear Polarized versus Random Polarized HeNe Laser

    There are two basic types of HeNe lasers with respect to polarization of the output beam: Random polarized and linearly polarized. With just a pair of mirrors and HeNe gas, the result will usually random polarized. To construct a linearly polarized HeNe laser, a Brewster-angle plate is added inside the laser cavity. (Search for: "Brewster Angle".) This introduces an orientation dependent loss with respect to polarization and virtually 100 percent of the photons will then be linearly polarized with the orientation with minimal loss. The Brewster plate on the types of HeNe laser tubes used here are buried within one of the mirror mount stems and are thus not readily visible.

    Although the experiments are described with a linear polarized HeNe laser, it is possible to use a random polarized laser instead if one is more readily available. However, its output will need to be converted to linear polarization in a very specific way with a linear polarizer to behave the same. But a random polarized laser can also demonstrate some interesting interferometer behavior not available with a linear polarized laser.

    Most random polarized lasers do not actually have polarization varying, well, at random. :) The term "random polarized" with respect to to HeNe laser simply means that nothing special is done to control the polarization. (I.e., no Brewster plate.) In the case of many red (633 nm) HeNe lasers, that means:

    1. There are two polarization axes instead of one.
    2. They are orthogonal to each-other.
    3. Their orientation cannot be predicted without testing but remains fixed for the life of the tube.
    4. Adjacent longitudinal modes have orthogonal polarization.

    And note that this only applies to red HeNes, not even other color HeNes, let alone most other lasers. Murphy must have taken a day off when the HeNe laser was invented because these attributes end up being quite useful - in fact fundamentally important - for many applications.

    A laser tube with these characteristics would be considered "well behaved". if the longitudinal modes move smoothly through the neon gain curve without abrupt changes in amplitude and not all are like this. :( :) A "flipper" will have "polarization switching" whereby at some point or points during mode sweep, the two sets of longitudinal modes will swap their polarization axes, usually instantly. For some truly nasty tubes, this will happen continuously, somewhat well, at random. Flippers are often not suitable for an interferometer because in the region of the flip, there may be excessive optical noise, possibly due to even the smallest amount of back-reflection, which will show up as large oscillations in the fringe signal. (While that in itself may be interesting, it will also be confusing.)

    The simplest way to test for a well behaved random polarized laser is to put the output through a linear polarizer and monitor it on a graphing laser power meter or photodiode and data acquisition system. Adjust the orientation of the polarizer for the maximum amplitude of the mode sweep variation. That aligns it with one of the polarization axes. Then inspect the plot over a couple minutes (from a cold start to get the fastest mode sweep) for abrup changes in amplitude. There should be none.

    The following animation shows the mode sweep of a random polarized HeNe laser similar to the JDSU 1107 or 1108. The red and blue lines represent the amplitudes of the orthogonal polarized outputs. To actually view these live with a similar display requires an instrument called a Scanning Fabry-Perot Interferometer (SFPI) with a dual polarization detector. While commercial SFPIs cost several thousand dollars, an SFPI with these capabilities can be built as a nice student project at modest cost. It's all done with mirrors. ;-)

    Mode Sweep of Short Random Polarized HeNe Laser

    A linearly polarized HeNe laser of similar length would have both modes be the same polarization and same color. :)

    The rate at which the modes pass through the neon gain curve will depend on how fast the tube is expanding from heating of the gas discharge, so it will slow down as it reaches thermal equilibrium from a cold start.

    Plots of the two polarized modes for a well behaved tube (non-flipper) would look something like:

    Mode Amplitude Plots of Well Behaved Short Random Polarized HeNe Laser

    The plots cover the time range from a cold start to close to thermal equilibrium. Note how both the red and blue plots are continuous. A full mode sweep cycle at the start is a few seconds while at the end it is a few minutes. After that it would be irregular as just ambient air moving around will have a significant effect.

    (For a linearly polarized tube with similar physical characteristics, the amplitude of the output would be the sum of the red and blue plots.)

    The plots of a typical flipper might look like the following (zoomed in to a few mode sweep cycles to show details):

    Mode Amplitude of Short Random Polarized "Flipper" HeNe Laser

    (The shape of the curves differ due to the tube not being the same model.) The vertical green line is the instant of the flip, which occurs quite close to the same location during each mode sweep cycle. However, experience shows that in the interferometer, there may be nasty stuff going on around that region and it won't be confined to an instantaneous event. The detected signal may be very noisy.

    For an academic challenge, a random polarized flipper is probably the most interesting type of tube to study. But a good understanding of what's going on is necessary to not to go insane attempting to decipher the behavior.

    The next most interesting tube would be one that is random polarized and well behaved. One or two of the kits put together to date includes that type of laser. A random polarized flipper can be provided in place of or in addition to the linear polarized or random polarized well behaved tube if interested. :-)



  • Back to Sam's Educational Michelson Homodyne Interferometer Project Manual Table of Contents.

    Homodyne V2.1 Summary of Features and Options

    Here is assorted information on what is for sale or will be for sale, directly from me or through eBay. V2.1 is virtually identical to V2.0 except for the size of some posts and spacers. V2.0 is functionally virtually identical to V1.0 and V1.5 except for the sizes of the breadboards, but a few other parts have changed, and thus the assembly procedures differ.

    There are three versions of the V2.1 Homodyne setups, though customization is possible.

    The Basic version really has enough for an introduction to interferometry and can be upgraded to the Deluxe version by adding the appropriate parts from me or elsewhere.

    Other changes compared to V1.0 and V1.5 include:

    1. V2.1 (and V2.0) use an aluminum plate on Thorlab posts rather than a wood block for the PBS and turning mirror mount as in V1.0 and V1.5. Both are very stable, but the plate affair is more in keeping with optical setups and doesn't contribute to the demise of trees. ;-) The wood block also had issues of degradation with multiple cycles of assembly and disassembly.

    2. A custom detector made from discrete parts replaces the Thorlabs DET110 in V1.0 and V1.5. The performance for the purposes of these kits is identical. And it facilitates adding a second channel for the quad-sin-cos decoder or X/Y detector and also simplifies the wiring with a terminal block and/or jumper wires instead of requiring BNC cables and clip leads. Both a solderless breadboard and assembled PCB version will be included.

      However, for the die-hard, a few DET110s are available at a lower cost than the current version of a biased photodiode from Thorlabs.

    3. The adjustable mounts will be either Thorlabs KM100s or Newport U100s. For the purposes of these kits, they are functionally equivalent and the Thorlabs are smaller and sleeker. ;-) The U100s are more precise (and a lot pricier if bought new!), though these versions lack knobs, requiring the use of a 5/64" ball driver, which may be less convenient. But never fear, appropriate Thorlabs knobs will be included in the kit.

    4. The QWPS and mounted CC in the Deluxe and Deluxe+ versions may be custom parts rather than from HP/Agilent. They will be functionally equivalent without issues of availability.

    5. A generic linear stage will be used, though a Parker 3902 may be an option. The generic stage is equally good for positioning here, possibly even more stable due to its wider track spacing.

    6. All versions will have either a Thorlabs RC carrier with RLA rail, or MR15 rail with MGN15-C block.

    Photos of each of the three versions with typical interferometers are shown below. The first is the Basic version with the 6x18" the breadboard, configured with Linear Interferometer and quad decoder PCB attached to the SG-µMD0 PCB:

    The next one is the Deluxe version configured for the High Stability Plane Mirror Interferometer. The laser includes a beam expander but it is otherwise configured the same way.

    The next is the Deluxe+ "Stretch" version on the 8x24" breadboard.



  • Back to Sam's Educational Michelson Homodyne Interferometer Project Manual Table of Contents.

    Sam's Educational Michelson Interferometer Interferometer Kit Parts List

    These are the parts for the Basic, Deluxe, and Deluxe+ setups except as noted.

    Replicating it should be straightforward

    The 8x18" optical breadboard for V1.0 and V1.5 is custom with the dimensions selected to be convenient for the projects while being able to easily ship Worldwide. The 6x18" and 8x24" breadboards for V2.0 and V2.1 are standard. If machining one, fewer than half the standard holes are enough based on all reasonable mounting locations, but that's only worth it if you're paying by the hole. ;-)

    Most other standard opto-mechanical parts are from Thorlabs. The major exceptions are the HP/Agilent PBS cube, QWPs, and mounted CCs (via eBay). But low cost substitutes for those are in the works.

    The JDSU 1107P and 1108P, or similar laser heads from Melles Griot/Pacific Lasertec like the 05-LHR/P-211 are ideal for these experiments. The criteria include being linearly polarized, eye-safe output power, and physical size. Optics companies like Edmunds, Newport, and Thorlabs have suitable lasers (though most are made by JDSU and PLT), but eBay is often a good source at a fraction of the cost. The laser could also be a bare tube safely enclosed, though the preferred linearly polarized variety is not that common at the low power of 0.5 to 1 mW. Random polarized heads and tubes can be used (and actually add some interesting areas to study) if certain criteria are met, primarily that they are not "flippers". See section: Linear Polarized versus Random Polarized HeNe Laser.

    Parts List for V2.1

    Differences between the versions will be noted.

    Parts in Homodyne Kits ONLY

    Quantity Description ------------------------------------------------------------------------------ Baseplate/Optical breadboard: 1x Aluminum optical breadboard 6x18" (Basic and Deluxe versions, BASE Lab Tools SAB0618) or 8x24" (Deluxe+ version, BASE Lab Tools SAB0824). Laser Assembly: 1x Laser consisting of 0.5-1 mW JDSU 1107/P or 1108/P, Melles Griot 05-LHP-211, or similar linearly polarized laser head and power supply with DC input brick and universal wall adapter, which may be used Worldwide. Minimum power 0.4 mW. A lab-style HeNe laser power supply may be an option. A random polarized laser is also an option. 1x (Random polarized lasers only): ~1x1" piece of circular polarizer. Consists of linear polarizer (+/-45 degrees) with QWP (0/90 degrees). Cut up as needed. 1x Beam Expander (Deluxe/Deluxe+ only) consisting of: 1x HP/Agilent 6 mm beam expander, modified, must be glued or press-fit into Beam Expander Adapter Plate. Resulting beam is 3-4 mm in diameter. 1x Beam Expander Adapter Plate. The front bezel of the laser has had 3 tapped holes added for this, or it can be glued. 3x M2.5 x 5 mm cap-head screws to secure adapter plate if not glued. 2x Laser Mounts each consisting of: 1x New or Old Small Ring Mount with four 8-32 Nylon thumbscrews. 1x Offset Plate with two 8-3/2 x 1/2" cap-head screw (New only). 1x Thorlabs TR3 post with 1/2" spacer and 8-32 x 3/4" cap-head screw to secure it to the offset plate (New Small 4-Screw Rings) or TR2 post and 8-32 x 1/2" cap-head screw (Old Small 4-Screw Rings). Custom Single Channel Detector and Quadrature Decoder: 1x Variable attenuator plate to be used as NPBS. 2x 1/2" piece of sheet (may be used as CP or LP), or CP+LP, cut to size. 2x Silicon photodiodes + spare. 1x 1K ohm resistor (PD protection). 2X Load resistors, 100K to 1M typical, or 1M trim-pot 1x Set of jumper wires, etc. 1x Solderless breadboard with adhesive back, 170 tie points. 1x Detector adapter plate with 8-32 or 1/4-20 1/2" setscrew. 1x QD1 PCB (option) with 8-32 1/2" cap-head screw and 1/4" x 1/4" #8 spacer. 2x 4 pole screw terminal block. 1x 1K ohm resistor. 1x 0.1 µF capacitor. 2x Photodiode. 2x 2 pin male to female socket strip. AB2 quadrature decoder parts kit. Custom Single Channel Detector and Quadrature Decoder: 1x Variable attenuator plate to be used as NPBS. 2x 1/2" piece of sheet (may be used as CP or LP), or CP+LP, cut to size. 2x Silicon photodiodes + spare. 1x 1K ohm resistor (PD protection). 2X Load resistors, 100K to 1M typical, or 1M trim-pot 1x Set of jumper wires, etc. 1x Solderless breadboard with adhesive back, 170 tie points. 1x Detector adapter plate with 8-32 or 1/4-20 1/2" setscrew. 1x QD1 PCB (option) with 8-32 1/2" cap-head screw and 1/4" x 1/4" #8 spacer. 2x 4 pole screw terminal block. 1x 1K ohm resistor. 1x 0.1 µF capacitor. 2x Photodiode. 2x 2 pin male to female socket strip. Common to both detectors: 1x Thorlabs PH2 post holder. 1x Thorlabs TR2 post (V2.1), or TR1 or TR1.5 post (V1.75). 1x BA1 or BA1S with 1/4-20 3/8" cap-head screw to attach to PH2 and 1/4-20 5/8" cap-head setscrew to attach to base. Micro Measurement Display 0 (µMD0): 1x Atmega 328 Nano 3.0 microcomputer board with pins soldered. 1x Homodyne firmware for Nano (may be preloaded or download from µMD0 Manual). 1x µMD Graphics User Interface (download from µMD0 Manual). 1x Solderless breadboard 3-1/4" x 2-1/4", 25 columns, 400 tie points. 1x LM393P dual voltage comparator. 1x 3 mm Red LED and 1-2K ohm resistor. 1x 3 mm green LED and 27-47K ohm resistor. 2x 100K to 1M ohm resistors or trim-pots. 2x 10K ohm trim-pots. 2x 100K ohm resistor. 2x 470K ohm resistors. 1x SG-µMD0 PCB (option). 1x 30 pin Socket for Nano (may need trimming). 2x 8 pin sockets for LM393 and UA9637. 1x 3 mm blue LED. 1x 10K ohm resistor. Support (Deluxe and Deluxe+): 1x 10K ohm potentiometer wired with 10K ohm current limiting resistor. - Hookup wire and jumper wires.

    Parts Common to both Heterodyne and Homodyne Kits

      Biased Photodiode Detector (BPD1):
    
        1x 1/2x1/2" piece of CP polarizer sheet (may be used as CP or LP with
           LP axis at 45 degrees), to be cut to size.
    
        1x 1/2x1/2" piece of LP polarizer sheet (LP axis XY), cut to size.
     
        1x BPD1 PCB with 8-32 1/2" cap-head screw and 1/4" #8 spacer.
        1x Silicon photodiodes + spare.
        1x 2 pin male to female socket strip for photodiode.
        1x 1K ohm resistor (PD protection).
        1x 0.1 µF capacitor.
        1X Load resistors, 2-3K ohms typical, or 10K trim-pot
        1x 4 pole screw terminal block.
        1x Solderless breadboard with adhesive back, 170 tie points.
        1x Set of jumper wires, etc.
    
      Common to BPD1, OR3, and AB2:
    
        1x Thorlabs PH2 post holder.
        1x Thorlabs TR3 post (V2.1) or TR2 post (V1.75).
    
        1x BA1 or BA1S with 1/4-20 3/8" cap-head screw to attach to PH2 and
           1/4-20 1/2" cap-head setscrew to attach to base. 
    
      Interferometer Assembly. (HP part numbers are used here; The actual parts
       may be the physically and functionally equivalent from Excel or Zygo.)
    
        1x HP/Agilent 10702A or 10706A PBS cube in frame with 4x 4-40 x 1-3/4"
           cap-head screws.
    
        1x HP/Agilent 10703A Cube Corner or mounted equivalent with 2 4-40 cap
           head screws.
    
        2x HP/Agilent 10722A Quarter WavePlate or custom equivalent each with two
           4-40 cap-head screws.
    
        1x Turning mirror (Approximately 1/2" x 1"). 
    
        1x Turning Mirror Bracket V2.0 with 4-40 x 5/16" cap-head screw and washer.
    
        1x (V2.1) PBS Mount Adapter Plate V2.1 with 2x Thorlabs TR3 post, 2x 3/4"
           spacer, and 2x 8-32 x 1" or 1-1/8" cap-head screw.
    
        1x (V1.75) PBS Mount Adapter Plate V2.1 with 2x Thorlabs TR2 post, 2x 1/4"
           spacer, and 2x 8-32 x 1/2" or 5/8" cap-head screw.
    
      Arm 1:
    
        1x Thorlabs KM100 or Newport U100 mirror mount with 3/8" 8-32 cap-head
            screw to secure it to post.
    
        1x 1" bare Cube Corner.
        1x 1" diameter planar mirror.
        1x Thorlabs PH2 post holder.  1/4-20 1/2" setscrew (V1.75 only).
        1x Thorlabs TR2 post (V2.1), or TR1 or TR1.5 (V1.75).
    
        1x (V1.75 only) BA1 or BA1S with 1/4-20 3/8" cap-head screw to attach
           to PH2 and 1/4-20 1/2" cap-head sscrew to attach to base.
    
      Arm 2:
    
        1x Thorlabs KM100 or Newport U100 mirror mount with 8-32 cap-head
           screw to secure it to post.
    
        1x 1" bare Cube Corner.
        1x 1" diameter planar mirror.
        1x Thorlabs PH2 post holder (rail or slide) and/or PH1 (MCPs or V1.75)
        1x Thorlabs TR1 post (rail or slide or MCP Type 1) or TR075 (MCP Type 2
           or V1.75).
    
        1x Generic 40x40 mm linear stage, end or side micrometer adjuster,
           maximum of 20 mm thick.  A 4-40 or M3 screw into a tapped hole is
           used to secure the post holder the top of the stage.
    
      Thorlabs rail assembly (Homodyne only):
    
        1x Thorlabs RC1 carrier with screw and washer to secure it to the
           bottom of the stage.
    
        1x Thorlabs RLA0600 6 inch rail (Basic or Deluxe version) or RLA1200
           12 inch rail (Deluxe+ version).
    
      Ball bearing rail assembly (Heterodyne):
    
        1x MGN15C carriage block.
    
        1x MGN15 stage adapter plate with 4x 4-40-3/16" cap-head screws to
           secure generic stage to plate and 4x M3-8 flathead screws to
           secure plate to MGN15C carriage block.
     
        1x Locking bracket with 2x M2.5x16 mm screws to secure bracket to
           MGN15C carriage block and 4-40 1/2" Nylon thumbscrew.
    
        1x Rail:
    
           1x MR15-150 150 mm rail with 2x 6-32 3/8" cap-head screws to secure
              rail to breadboard and rubber bumpers to be used as end-stops.
              Breadboard requires addition of two 6-32 tapped holes. (Compact
              version.)
    
           1x MR15-200 200 mm rail with 2x 6-32 7/8" cap-head screws and nuts to
              secure rail to breadboard.  The screws with spacers also can act
              as the end-stops (Mid-Size version.)
    
           1x MR15-300 300 mm rail with 2x 6-32 7/8" cap-head screws and nuts to
              secure rail to breadboard.  The screws with spacers also can act
              as the end-stops (Extended version.)
    
      Motion Control Platform (Option):
    
        1x Mini-stepper driven platform, 100 mm, 150 mm or 200 mm travel, with
           4x 4-40 1/2" cap head screws or 4x 4-40 3/4" cap-head screws
           and nuts to secure it to the breadboard.  Each of these will nearly
           entirely fit on the breadboard for the Compact, Mid-Size,
           and Extended Versions, respectively.  The hole pattern may differ
           for the Type 1 and Type 2 MCPs.
    
        1x MCP adapter plate (if not already present with the generic stage, they
           are the same.
    
        1x STEPPERONLINE DM320T micro-step stepper motor controller.
    
        1x Atmega 328P Nano 3.0 microcomputer board with solderless breadboard.
    
        1x Optical high resolution rotary encoder to move the platform via
           the Nano at least for testing.
    
      Voice Coil Actuator:
    
        1x 1-1/2" to 2" loudspeaker. 
        1x 1" D x 1/4" T aluminum or Acrylic Speaker mounting Disk.
        1x Speaker mirror (Approximately 1/2"x 1", may be same as turning mirror).
    
      Piezo Transducer (Deluxe):
    
        1x 27 mm PZT beeper element.
        1x #2 washer to use as spacer.
        1x PZT mirror (Approximately 1/2"x 1", similar to turning mirror).
    
      Gas Cell Compensator (Air Pressure and Temperature):
        
        1x 1" OD, 7/8" ID, 2" L Acrylic tube.
        2x 1-1/8" D glass or Acrylic window.
        1x 10-32 to hose barb adapter.
        1x 8-32 3/8" or 1/2" set-screw.
        1x Blood pressure bulb with valve.
        1x Blood pressure gauge.
        1x 3-3.5 mm (or 1/8") ID Rubbor tubing to connect.
        1x Hose barb "T" for 3 to 4 mm tubing.
    
      Thermal Expansion:
    
        1x  ~1x1x2 cm compensator plate or other glass block with polished sides.
        1x  25 ohm, 10 W power resistor.
        1x  12 VDC 1 A power pack (PD bias, voice coil, PZT, thermal expansion).
        1x  Screw terminal to 5.5/2.1 female barrel connector adapter.
        1x  ~1" x 1-1/4" x 1/4" Delrin "chip" with 8-32 set-screw.
    
      Parts common to Gas Cell Compensator and Thermal Expansion:
    
        1x Thorlabs PH2 post holder.
        1x Thorlabs TR3 post (V2.1) or TR2 post (V1.75).
    
        1x BA1 or BA1S with 1/4-20 3/8" cap-head screw to attach to PH2 and 1/4-20
           1/2" cap-head setscrew to attach to base. 
    
        1x Two part 5 minute Epoxy.
    


  • Back to Sam's Educational Michelson Homodyne Interferometer Project Manual Table of Contents.