Table of Contents



Introduction

This note outlines the procedure for replacing the glass HeNe laser tube in an HP/Agilent 5517 or 5501B laser.

This isn't rocket science, but a high degree of attention to detail and care is required to assure a successful outcome. Familiarity with metrology lasers is not necessary, but if you don't know anything about HeNe lasers in general, then gaining some experience with their care and feeding may be desirable prior to tackling the replacement, if for no other reason than to avoid shocking experiences and destruction of expensive replacement tubes or expensive lasers. In addition, some minimal electronics experience is required to be able to perform certain tests and make final adjustments.

A note about the formatting of this document: Clicking on most of the photos will bring up a high resolution version. To be able to view this along with the text, it is recommended that your browser be set to open a new window rather than a new tab (as may be the default with Firefox). They will all appear in the same (new) browser window so there is no chance of cluttering up your desktop.

Safety

There are two relevant areas: electrical safety and laser safety. HP/Agilent 5517 and 5501B lasers run on +/-15 VDC at a maximum current of less than 2.5 amps. So, basic common-sense precautions will suffice. However, the HeNe laser tube requires around 1,600 V at 3.5 mA, provided by a DC-DC converter "brick" power supply. Its output is not particularly dangerous, but touching it may result in a reflex reaction tossing the $1,000 glass laser tube across the room. Even after powering off, a painful charge remains on the power supply capacitors for quite awhile. Once the new tube has been terminated with the cathode/heater wires and anode wire, and everything is properly insulated, the risk is minimal. But if in doubt, use a jumper wire AFTER POWERING OFF to briefly short between the tube anode (connected to the fat white wire, center contact of the black HV connector) and the red or violet cathode/heater wires. DO NOT short directly to the metal chassis as there's a chance this could result in destructive current flow through the control PCB.

The maximum optical output power from the laser tube itself for any of these lasers is well under 1 mW. It still may be bright but is not likely to result in permanent vision damage from a momentary exposure. However, this should be avoided as the after-images may remain for a long time. Always view the beam projected onto a white card or similar screen. There should never be any need to stare into the beam, even with your remaining good eye!

Required Tools, Test Equipment, and Materials

In addition to these, some home-built/custom items will be required:

The HP/Agilent 5517 or 5501B Metrology Laser

These are red (~633 nm wavelength) HeNe lasers producing two optical frequencies differing by anywhere from 1.5 to over 7 MHz depending on model. The two frequencies are denoted F1 (lower frequency) and F2 (higher frequency). For all the 5517s (and 5518A and 5519A/B), F1 is horizontally polarized and F2 is vertically polarized. For reasons probably no longer known to anyone on the Planet, these were reversed for the 5501B and earlier HP metrology lasers. The exact wavelength has 6 more significant figures but for our purposes they are both red. :) And since there is nothing that can be done to alter the optical frequency once the tube is rebuilt, it's not something that is worth measuring. However, once the laser is put back into service, a full dimensional calibration will be required as there is no guarantee that a rebuilt tube will operate at exactly the same wavelength as the original HP/Agilent tube to all those significant figures. Even so, environmental factors like temperature, humidity, and pressure have a more significant effect on measurement calibration than any conceivable change in optical frequency.

     

Typical HP-5517D Metrology Laser

The general appearance of all the 5517 and 5501B lasers is similar, but those made by Agilent after around the year 2000 have a cheesy metal shroud in place of the 2-piece gray cover. The 5517B, 5517C, and older 5517Ds, as well as all 5501Bs have a physically similar tube assembly with a "long" glass laser tube. The procedure below is written for those. Some newer 5517Ds, and the much less common 5517E/F/G lasers use a "short" glass tube, but are otherwise also very similar. But minor modifications to the procedure will be required for those if the time comes.

The first step in the tube extraction process is to remove the laser's feet (if present) and laser cover. The feet are removed since accurate positioning of the laser for checking and setting beam alignment requires the screw holes on the baseplate of the laser. It would not be precise enough with the feet present.

The feet simply unscrew, but on some lasers, they are quite tight. Clamping a recalcitrant foot in a bench vice and rocking it back and forth is usually is effective to loosen it.

If there is a shutter wheel on the front of the laser, rotate it so the large hole is at the bottom. The flat-head screw engages a quarter-turn fastener. Use a flat-blade screwdriver that just fits the head and press it in and turn it counter-clockwise 1/4 turn. The screw should pop out (but remains captive) and the front plate will come off. Then, where there are covers as above, they can be spread apart to remove. On the newer lasers, there is a rubber gasket of sorts all around the metal shroud, so some force may be required to push in the front plate to allow the screw to turn. Then spread the bottom apart slightly and slide it off. Take care to not get it hung up on any components on the control PCB. For a pre-2004 laser, the result should be very similar to the photos below.

     

HP-5517D with its Cover Removed

Most newer Agilent use a control PCB with mostly surface mount parts but it essentially emulates the older one so very little changes as far as tube replacement is concerned. A very few 5517s have a third type of PCB. And some newer lasers have an additional sheet metal cover inside that necessitates a different tube assembly removal process. Deal with those on a case-by-case basis.

Testing the Laser Prior to Disassembly

Note: It is assumed that the laser tube has been determined to be end-of-life and too weak to be useful. Thus, no actual testing of its performance is provided here. If in doubt, test the tube and/or the entire laser to confirm that the tube needs replacement and it's not just a blown fuse!

Set up the laser on the Laser Alignment Table (LAT) so that it is in the "Standard Position". Power up the laser. As long as a beam appears, no matter how weak and no matter how long it takes to appear, or even if the beam sputters or flickers, it will be possible to check alignment. The center of the beam should be within +/-1 mm of the marked position on the LAT screen for the 5517B/C and most 5517D lasers. There are some versions that shift it by about 1 mm for reasons unknown. If this laser is like that, mark the new position so when the tube is replaced, it can be aligned to that mark.

Removing the Laser Assembly

The physical aspects of replacing and aligning the laser tube and optics are quite straightforward in principle but the details can get messy, especially with respect to extracting the glass tube from the magnet assembly. Until a better technique is found than the one below, that will represent the most single time consuming portion of the overall process.

The next step is to remove the tube assembly from the laser. A typical sample is shown below. Some details may be different on yours like the method of securing the magnet assembly (the 3 long screws) and the style of the beam expander. But in general, it will be almost identical in all other respects. On the left are the high voltage (HV) connector and 2 pin cathode/heater connectors. The back-end of the glass tube is sticking out the back. The large shiny object is the Zeeman magnet. The black beam expander and waveplate assemblies are at the right.


Typical HP/Agilent 5517B/C/D and 5501B Laser Tube Assembly

Removal requires cutting any cable ties, disconnecting the black high voltage connector and unplugging the 2 pin cathode/heater connector, and taking out the 4 screws below the baseplate. It should then be possible to maneuver the tube assembly out of the laser body.

Note: If there is a plastic cover on the back of the tube assembly, it has a "short" tube. The procedures to replace it differ in minor details but should be similar to those for the "long" tube assumed here.

Availability of Replacement HeNe Laser Tubes

To be usable as a replacement for the custom HP/Agilent, it must satisfy very specific optical, electrical, and physical requirements, as well as having a life expectancy that makes the entire rebuild process cost effective. At this time, there are no standard or even semi-custom commercially available tubes that even come close in either performance or longevity. While there are a hand-full of third party companies claiming to be able to install non-HP/Agilent tubes, these often have serious problems in both departments. Even if the output power and REF frequency are acceptable, there may be rogue modes, a beam size/profile that makes operation with standard tools tricky or impossible, and short life. They will also take longer to lock, and may be more sensitive to temperature. Therefore, for the remainder of this document, it is assumed that an identical HP/Agilent tube that has been regased or totally refurbished will be used as the replacement. The term "regased" means that the tube was simply refilled with pure He and Ne in the appropriate ratio and then sealed. The term "refurbished" means that in addition to regassing, the tube may have been totally disassembled so that parts like the cathode and mirrors could be treated or replaced. Depending on the condition of the old tube, a complete refurbishment process may be required to achieve adequate performance and life.

At present, only one company is known to be willing to do at least a regas on these tubes, and then only if the glass tube is removed from the magnet assembly. However, our first test with them is not yet entirely satisfactory as the tube came back with either an external leak or internal contamination.

Furthermore, unless the same gas fill (He:Ne isotopes, ratio, and pressure) are the same in the rebuilt tube, both the optical frequency (and thus wavelength) and REF frequency will change. In particular, the use of the natural isotope ratio for He and Ne rather than an enhances one as is believed to be used by HP/Agilent will result in a wider neon gain curve as well as shift in the peak of the neon gain curve. The same will occur with a higher fill pressure. A dimensional calibration will take care of any optical frequency chnage. But REF may very well end up out of spec (usually lower) for the particular model laser. And there's also a chance that rogue modes will show up due to the wider neon gain curve. This would necessitate a reduction in the magnetic field to eliminate (or at least minimize) them, further decreasing REF. So these possible side effects should be anticipated. For example, regassing a 5517D resulted in REF dropping to under 3.0 MHz. (Spec is 3.4 to 4.0 MHz.) Part of this was due to the lock point not being where REF peaked (which is where it usually is on HP/Agilent lasers) even though the F1/F2 components were balanced. The laser would then not be acceptable for service where a 5517D were required, but could be converted to a 5517C. Increasing REF might be possible with a stronger magnet or by adding magnets, but this would be very likely to result in rogue modes even if none were present originally. Be prepared with a fall-back plan!

Initial Disassembly of Laser Tube Assembly

Our ultimate objective is to break it down (no pun intended...) to its component parts without literally breaking anything:


Components of a Typical HP/Agilent 5517B/C/D and 5501B Laser Tube Assembly

(The exact style of the beam expander and some other parts may differ slightly depending on the model and age of the laser.)

Only the waveplate assembly and beam expander can be removed easily:

  1. Put marks or labels on the top front of the beam expander and waveplate assemblies so that they can be replaced later with the same orientation.

  2. Remove the three screws securing the waveplate assembly and put it in a clean plastic bag to prevent contamination. DO NOT touch or attempt to clean the optical surfaces as they will be damaged. DO NOT put the screws in the same bag as they can damage optics!

  3. Remove the three screws securing the beam expander and do the same. The beam expander may be slightly stuck to the aluminum housing but should pop free with just a bit of hand persuasion.

  4. Measure the precise distance that the front of the glass laser tube extends beyond the beam expander mounting surface. When the tube is replaced, this should be matched as precisely as possible to assure that the collimation will be acceptable. Some adjustment may be possible but best to avoid having to do that.

Removing the Laser Tube Intact

The main difficulty in the entire process is with respect to the rubbery potting compound used to secure the tube inside the magnet assembly. While the stuff is relatively soft, I am not aware of any solvent that will dissolve it without also eating the metallic parts of the tube and its surrounding structure, as well as rotting human internal organs (if there are any at all, solvents that is). The potting compound appears to be some sort of expanded RTV-like material which is impervious to modest heat as well - it may just get harder at high temperatures. Furthermore, much of the potting compound that needs to be removed is between the tube, and the magnet and the aluminum front section, which is a gap of only around 2 mm, which makes it tough to get a solvent into those areas.

Therefore, the only known methods of extracting the tube all require some form of mechanical removal of the potting compound or total discombobulation of the magnet and possibly the front section as well. In principle, the easiest technique might be to slit the magnet and front section lengthwise with a high speed low vibration diamond cutoff wheel and then spread them apart and remove the two halves. With care, the remaining potting compound is then easily picked off using a sharp blade and other common tools. To install the refurbished tube, a commercial magnet could replace the destroyed HP/Agilent magnet, and the front section could be reconstructed with shims or hard Epoxy to fill the cutoff wheel kurf. What is not known at this point is (1) whether the glass tubes would survive consistently using this approach and (2) a supplier of replacement magnets (but there should be many).

Another alternative that has been suggested is to use a high pressure water jet. Whether this would be able to remove the potting compound deep inside the space between the tube and magnet is not known.

Thus the technique I have used to do a half dozen or so of these is to cut and pick away the potting compound using a variety of tools normally associated with torture - dental picks, long thin hex drivers, thin pieces of sheet metal, coping saw blades, and so forth. These are all used to gradually eat away at the potting compound at the back and in the space between the glass tube and magnet. This requires on average about three hours start to finish. As a practical matter, even trained monkeys promised unlimited bananas would quit after doing one or two of these. :( :) There has to be a better way. But here goes assuming the tube assembly has been removed from the laser and the waveplates and beam expander have already been taken off:

Note: Despite the extra effort, it may be worth practicing the following first on tube assemblies that have no rebuild value, or at least ones where a truckload is available. :) Then it won't be so bad to accidentally smash a few.

  1. Remove the screws securing the feet brackets and front section to the magnet. (Three or six places.) Some of these may be quite tight.

  2. Remove the back feet bracket. It may be necessary to cut away a small amount of the potting compound around the wires to clear what's left.

  3. Using an Xacto knife or single edge razor blade, carefully cut away the potting compound at the back of the tube just to the point of exposing the connection points of the wiring to the tube terminals - 2 for the red and purple cathode/heater and the anode. If the tube is definitely going to be regassed or completely refurbished, the wires should be cut about 1/2 inch from to the terminals leaving just enough bare wire remaining to be able to solder during reassembly.

    Cut the wires about 1/2 to 1 inch from their termination to provide enough of a stub to reattach them after the tube has been regased or refurbished. DO NOT go any deeper at this time. Take photos of the rear of the tube and magnet showing the orientation of the cathode/heater and anode terminals. This will then enable the orientation to be matched during reassembly.

    Or, if you would like to keep the wiring intact, do NOT cut away more of the potting compound than necessary to remove the back feet bracket. Simply tape the heater/cathode and anode cables to the magnet and wrap the entire thing with some padding like thin packing material to protect them. This is less convenient but would be desirable if, for example, your intention is run the tube after removal. In this case, put a mark of some kind on the remaining potting at the rear of the tube to use as a guide to orientation upon reassembly.

The next step is to remove the aluminum front section.

  1. Carefully clamp the front section in a padded bench vice taking extreme care to avoid contacting the nose of the glass tube. Gently rock the magnet back and forth while twisting and issuing appropriate chants, incantations, and 4 letter words. :) With some luck, the front section may come free of the glass tube. Just remember that this is basically a fragile glass bottle inside a massive metal structure. So, don't get carried away! If there is any indication of it loosening, the continue and it may eventually come free.

    If this doesn't work, wedging a blade or two between the front feet bracket and front section to attempt to lever the front section off may work but again don't get carried away.

    It may be necessary to pick away at the potting compound accessible through the fill port on the top or side, or even from the front to loosen it up.

    Going back and forth between this and the approaches above may eventually yield results. But if this doesn't seem to be going anywhere, it may be necessary to cut the front section apart, or at least cut off the front of the front section to gain better access to the space between the glass tube and aluminum housing. It can always be reconstructed later.

Assuming the front section is removed intact, some of the rubbery potting material will probably remain inside. This is actually preferred as it will serve to center the replacement tube when the time comes. So, only remove loose bits or pieces about to fall off.

Once the front section has been pulled off, it's time for the main event - removal of the potting compound from between the tube and magnet.

  1. Use an Xacto knife or single edge razor blade to slice away the potting compound at the back of the tube so that what remains is a cylinder about 5-7 mm smaller in diameter than the inside of the magnet. DO NOT make it any smaller as the remaining potting compound will serve to protect the remains of the fragile glass tip-off and feed-throughs during the tube removal process.

  2. Now if you weren't having fun up to this point, here's your chance. The idea is to get in between the tube and magnet and pick away at the potting compound one bit at a time. This is by far the most time consuming and boring part of this entire operation. My principle tool is a 1/16th inch hex driver. Unless the tube is way off-center, this can get in in the confined space easily and with care, will not damage the tube. But don't force anything! Take your time. (I can hear the trained monkeys already complaining.)

    Work from both front and back but avoid going toward the center in the back to prevent damage to the fragile tip-off and feed-throughs. Didn't I say that already?

    (Note that if this is an older 5501B, then the magnet consists of 4 segments. Working from the front, it's much easier to go one segment at a time and then remove that magnet and set it aside. However, don't get carried away removing the last segment at the back. It is still possible to damage the glass tip-off or feed-throughs or break the tube! But all lasers made after about 1992 have one-piece magnets.)

  3. Once first contact is achieved - meeting of the excavation so that a thin tool can be pushed all the way through - it will go faster. Then, the idea is to work around the circumference until somewhat over 1/2 of the tube is free. A piano wire or coping saw blade may speed this process, though I haven't tried them. Eventually, with some wiggling, the entire tube should come free. It will have a lot of potting compound debris stuck to it but be intact.

  4. Use a single edge razor blade or Xacto knife to carefully peel away the remaining potting compound from all nooks and crannies of the tube. Be particularly cautious in the are of the anode, tip-off, and feed-throughs. It's still possible to screw this up!

  5. The result should look something like the photo below except perhaps that the wires are cut short.


    Typical HP/Agilent 5517B/C/D and 5501B Glass HeNe Laser Tube

Testing the Replacement Laser Tube

It's a good idea to run the replacement laser tube and measure its output power and mode behavior before installing it. While the chances of a new tube being defective are relatively small, it's been known to happen. And this will likely be one that has been regased or refurbished where any number of things can go wrong. In fact, one tube that was regased by a well-known company was barely lasing when it was returned. At first it appeared as though there was a leak - and that may still be the case. But after running for a couple days, it cleaned up to spec'd power and has been fine ever since (several months so far). My assumption is that there was an "internal" leak - some contamination that was adequately baked out trapped somewhere and it needed to be adsorbed by the cathode.

  1. Temporarily mount the replacement tube inside the magnet intended to be used, presumably the magnet from which it was extracted if that's how it was done. For this test orientation is not critical, but setting it at the original orientation (if known) would be preferable. Use insulating material to approximately center it in the magnet with the anode terminal lined up at the end of the magnet.

  2. Carefully attach the ballast to the anode lead wire. Connect the return to one of the cathode/heater wires (doesn't matter which one). These connections do not have to be soldered but should be solid enough that there won't be intermittent contact, which can damage the power supply and tube. Enclose the anode connection or the entire tube stem with an insulating tube or sheet to prevent arcs and short circuits.

  3. Power up the laser tube using the HeNe laser power supply and extension cable. A regased or refurbished tube should light almost instantly.

  4. Apply 5 or 6 CDC to the heater wires to accelerate mode sweep. This will enable the tube to reach an equilibrium temperature that is similar to that when it is locked.

  5. The output power will vary from a minimum to a maximum with a gradually increasing period. The variation may be 25 percent or even more. The normal lock point is generally at the minimum so that's the power level to check to determine if it is adequate. The average output power will gradually increase as the tube warms up by up to 25 percent or more, so wait till the mode sweep has slowed to a period of 20 or 30 seconds before testing for the minimum output power.

  6. The REF frequency can also be tested at this time. Place a linear polarizer in the output beam. Use the photodiode detector and oscilloscope or frequency counter to test it. Without the waveplates, the orientation of the polarizer does not matter. A beat signal will only appear for a fraction of the mode sweep - 10 to 25 percent depending on the model tube. The relevant value is the maximum frequency which will be near the minimum of the output power. If the tube/magnet combination is the same as it was originally, the REF should be comfortably within spec for the laser model. Preferably, it should be between 10 percent above the minimum and 10 percent below the maximum. Don't obsess if it isn't quite within this window, but it shouldn't be on the hairy edge.

    However, since the gas-fill may not match that of HP/Agilent, it's quite possible REF will differ substantially - probably lower - from the original spec for the tube model. A modest out of spec REF can be adjusted. But if it's way low, it may be neceesary to reclassify the model of the laser - a 5517D to a 5517C for example.

  7. To be as nearly 100 percent sure that the new tube will behave properly, running this setup on a data acquisition system monitoring the output power with a fast enough sampling rate to catch mode flips or ms-length dropouts might be desirable. However, if the only problem with the tube originally was low power or sputtering due to high dropout current, this may not be worth the effort.

Installing The High Voltage Wires

  1. Slip some pieces of high voltage heat-shrink over the red and purple cathode/heater wires and attach them to the proper terminals as shown in the above photo. The purple wire goes to the terminal that is connected to the heater and to the cathode can. The red wire goes only to the the heater. Make sure the connections are mechanically secure and then solder. Inspect the solder joints! Shrink the heat-shrink.

  2. Do the same with the anode wire taking extreme care not to flex the metal strip more than necessary to avoid breaking it off. Reattaching would require either a spot weld or mechanical screw clamp - it should NOT be soldered as the heat may crack the tube. The distance from the anode terminal strip to the start of the white wire covering should be as small as possible - no more than 1/4 inch - to minimize the area that needs to be insulated.

  3. Secure the anode wire to the glass stem with a pair of thin cable ties.

  4. Perform a quick test to assure that the tube is still happy. Then discharge the caps! ;-)

  5. Cover the exposed anode terminal and its white wire up to 1/2 inch past the start of its insulation with a generous layer of non-acidic RTV and allow it to cure overnight.

Installing the Replacement Tube in the Magnet Assembly

  1. If the front section was removed intact with significant potting compound remaining, it will serve to center the tube. Otherwise, if there is not enough remaining, and/or the housing had to be rebuilt, it might be desirable to drill and tap holes at 3 locations around the housing for Nylon thumb-screws to be used to temporarily center the tube until the new potting compound has been injected and cured.

  2. Reassemble the magnet, feet brackets, and front section as they were originally. The label is upright on the side with the tube facing to the left in case you forgot. ;-)

  3. Place the magnet assembly on the LAT in the Standard position.

  4. Install the glass tube inside the magnet assembly oriented as it was before removal if known, or with the anode terminal pointing up if not.

  5. Add rubber wedges or shims to center the back of the laser tube in the magnet.

  6. Power the tube and adjust the wedges or shims to center the output beam on the previously determined target.

  7. Check the position of the beam near the laser to confirm that it meets HP/Agilent specs for the particular laser model.

  8. Adjust the longitudinal tube position so it pokes out by exactly the same amount as before removal. If the beam expander is adjustable like the one in the photo of the intact laser tube assembly, above, then this isn't so critical as it can be tweeked later. But it if is like the one in the photo of the disassembled components, there is no way to adjust that, at least not easily.

  9. Inject a small amount of the non-acidic RTV at several places around the back of the tube to temporarily secure it.

  10. Install the beam expander in the same orientation it was in originally with the screws just the least bit snug. With the laser tube powered, adjust its position to obtain the most symmetric beam profile projected on a white card. Tighten the screws and recheck it.

    Optional: To have confirmation of the correct tube position, project the beam 50 feet away. A 6 mm beam should still be no more than 7 or 8 mm in diameter. Fine tune the longitudinal tube position to minimize its size if necessary. Yes, the beam will look rather ratty at that distance. :( :)

  11. Install the waveplate assembly in the same orientation it was in originally.

  12. The laser aside and allow the RTV to overnight. Repeat the alignment and collmation tests. Assuming they are satisfactory, inject RTV all around the rear of the tube as well as into the fill hole on the side of the front section.

Installing The Tube Assembly

  1. Reassemble the laser as it was originally. Install all 4 screws but only tighten the flat head screw(s) slightly and leave the others loose. Take care with the exposed glass stem at the back of the tube as it is somewhat more fragile now than when it was fully potted. Route the ballast behind the magnet if there is no enough slack to go under the tube assembly (or as appropriate depending on the original layout).

  2. Place the laser on the LAT and power it up. Confirm that the vertical alignment is still on the target. Adjust the tube from side to side to center it. Then tighten all 4 screws.

Initial Test

  1. Perform the temperature set-point adjustment if necessary. For a tube reinstalled in the same laser body with the same control PCB, this should not be needed, but checking it won't hurt.

  2. Power up the laser and allow it to warm up and lock. If the waveplates are in the same position as originally, there should be no problem locking in the normal time. However, fine tuning of the waveplates may still be required.

  3. Let it run for an hour or so to confirm that it is stable.

Testing for Single Longitudinal Mode Output

At this point, those who are anal retentive would want to test the output of the tube to assure that only the desired longitudinal mode is present. The presence of significant power in the second longitudinal mode that's lasing (but should be blocked) can potentially result in small positioning errors over distances on the order of 10s of cm. How large the output power in the second undesirable mode needs to be before this becomes an issue isn't known. And I've seen at least one almost new HP/Agilent laser that had one.

The test will require the use of a Scanning Fabry Perot Interferometer (SFPI, also known as a laser spectrum analyzer) to display the mode structure of the laser. As a practical matter, there isn't much that can be done if a rogue mode is found, so as long as it is small (2 percent or less) ignore it.


Scanning Fabry-Perot Interferometer Positioned in Front of Laser

While the photo shows the SFPI very close to the laser, it's better to Position it further away since alignment that results in the best display will result in a reflected beam directly back to the laser. There is the potential for that to destabilize the laser causing it to lose lock. Keeping it just a little bit off-axis will help greatly, and positioning the SFPI head further away will make this easier. A distance of a 2 or 3 feet would be more than adequate. Having said all that, it worked fine as pictured. :)

Adjust the SFPI mount so that the laser beam is centered on the SFPI input lens. For the height, use shims under the laser or SFPI mount if it is not adjustable.

For a typical SFPI like an SP-470 head attached to an SP-476 driver, attach the cables as follows. Adhere strictly to the color code if available.

The controller should be set for "Free Run", maximum sweep rate (Time fully CW), X1 Dispersion, with the Variable and Centering controls about mid-range.

The scope should have the Vertical Amp (green) going to Channel 1, Trigger Out (yellow) going to Ext Trigger or Channel 2 (with the scope set for triggering using Channel 2). Start with Vertical Gain second position from CCW and Variable centered. On the rear panel, the slide switch above the HV Scan Output should be set to low or 300 V and the Blanking Output control should be set fully CW. With these settings, there should be something on the scope if alignment of the SFPI head is reasonable. Fine tune alignment for maximum amplitude and adjust controls for appropriate horizontal and vertical display. Refer to the SP-476 Operation and Service Manual for more details.


Scanning Fabry-Perot Interferometer Display of Laser Output After Locking - ~8% MM, ~1% MM, Pure SM

The three photos shows varying degrees of rogue modes. The one on the left has a unwanted mode at about 8%. This would be really bad and unlikely to occur unless a tube susceptible to rogue modes were installed in a much stronger magnet than it was in originally. Or, an inappropriate non-HP/Agilent tube were used. The center one is about 1% and while not optimal, is probably acceptable. The one on the right is 100% pure SM, which is the way most HP/Agilent lasers are built.

Note that single peaks like those in the photos are a deception. There are actually a pair of peaks separated by the REF frequency, but they are too close together to be resolved by the SFPI.

Waveplate Adjustment

If the tube orientation is identical to the way it was originally, the waveplates should still be set correctly. If not, then the most likely effect will be for the F1/F2 orthogonality to be less than optimal. The following is easiest to perform on the ITA.

  1. Power up the laser on the ITA and allow it to come READY.

  2. Connect an analog multimeter or microammeter (50 µA full scale) between the test point on the 10780A/B/C optical receiver and its ground lug. The output is approximately 50 µA for maximum signal but is highly non-linear. With a rebuilt tube (or almost anything that's not dead!), the meter should read around 50µA.

  3. Place a linear polarizer in front of the laser set to be *exactly* horizontal. If the waveplates are correctly adjusted, the signal level should drop to near 0. This is very sensitive with respect to both waveplate and polarizer orientation.

  4. Using the waveplate adjustment tools, very slowly and carefully rotate the main barrel of the output (half) waveplate to minimize the measured signal level. The main barrel adjustment holes are the ones closest to the metal mounting plate. DO NOT touch the outer barrels - they almost never need adjustment. If it goes below 10 µA, that's probably good enough. But by alternately turning the inner barrels of both the input (quarter) and output (half) waveplates in very small increments (a few degrees), it may be possible to reduce it further.

  5. Once that is accomplished, the waveform from the optical receiver viewed on an oscilloscope should be relatively clean and change only in frequency if the reflector is moved repidly back and forth.

Final Testing, Adjustments, and Inspection

  1. Confirm that the laser current is 3.5 mA +/-0.1 mA. With power OFF, connect a 1k ohm resistor between the purple wire and the laser ground (chassis). Measure voltage across the resistor. Calibration is then 1 mA/V. (This is safer than installing a current meter.) With a healthy tube, a current reading out of spec almost certainly is caused by a defective brick or +15 VDC out of tolerance.

  2. Recheck alignment. There's not much that can be done easily for vertical alignment at this point except add shims, but horizontal alignment can be fine tuned by loosening the 4 screws securing the tube assembly.

  3. If desired, add an hour meter to allow for a deferred warranty.

  4. Install cable ties to secure the anode cable and ballast.

  5. Perform normal housekeeping to make the inside look spic and span. :)

  6. Inspect the optical window on the front plate and clean if necessary. Replace the cover.

  7. Power up the laser and allow it to run for several hours on a monitored setup like the ITA (5508A or 10887A/P). There should be no errors.

  8. Record the output power and REF frequency at power-up and after two hours.

YOU ARE DONE!

Extended Testing

Optional: Set up the laser with a polarizing beam splitter, dual photodiodes, and PC-based data acquisition system, and capture output at a minimum 50-100 samples per second for 24 hours or more. Inspect the data to assure that there are *NO* mode flips, glitches, or other unsightly blemishes in the laser's output power for either polarization after the laser has locked. This would catch infrequent (but fatal in terms of performance) problems in the HeNe laser power supply, ballast resistance, or wiring, problems with the control PCB; or a misbehaving tube. A recording laser power meter can also be used, but it will not likely have a high enough sample rate to capture short events. Here are the most likely symptoms and causes:

  1. Glitch (dip) in both polarizations: Possible laser tube current dropout due to high mileage tube, or high (increased from use) value or nearly open ballast resistor. Both of these would be very unusual.

  2. Loses lock and re-establishes: The temperature set-point is incorrect, defective control PCB, defective heater inside tube or bad connections to it, defective DC power suppl(ies).